Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Siemens Sequoıa Problem

(93)
(9)

kenan celik

Siemens Sequoıa Problem

March 18, 2008 06:59

when the start the machine a error giving,writing on the screen error 25,27,28
hit (cr) enter to service menu.
how is the service menu?
or what is the problem.

i mean a cmos ram clerad on the ico3 board because some body opened the system case.
now machine is not working.

i put the a othet machine ico3 board and hdd driver then system normaly worked.
how can load the software or what can i do????

Log inor Register

to rate and post a comment

Jim Carr

Re: Siemens Sequoıa Problem

March 18, 2008 04:57

If you removed the metal panel from the card cage without having "Third Party Service" enabled, then the only way to fix the system is to reload software or get a Siemens password to reset the system. In the United States and many other countries, Siemens will provide a software installer MO disk that is specific to your serial number.

Log inor Register

to rate and post a comment

(39)
(1)
(2)

Emerson Freire

Re: Siemens Sequoıa Problem

March 18, 2008 05:00

Dear Kenan,

You will have to call Siemens to fix this.
The Sequoia has sensors on the system case so if they are triggered you will need a password to reset them.
This password changes every day and only Siemens has it...
So all service has to be done by Siemens...

In order to access the service menu you will need this password and power up the system holding this two keys [multihertz up]+[s] (It will appear in the display Keys detected:15 b0).

This errors you have are:

25 - Test TX1 VGain Slot1
27 - Test PPS control status lines
28 - Test TX1 PWG registers Slot 1

All of them relate to the CN board....

I hope this helped...

Best Regards,

Emerson.

Log inor Register

to rate and post a comment

(120)
(6)
(1)

manuel valdez

Re: Siemens Sequoıa Problem

February 14, 2012 07:39

Dear Emerson i have the same problem i am from mexico how i can cantact to siemens? any email?
thanks and best regards

Log inor Register

to rate and post a comment

Fred Fimbres

Re: Siemens Sequoıa Problem

February 15, 2012 09:34

Manuel, buenos dias. Please contact me, we might be able to help you with this problem.

Saludos;

Fred Fimbres
UltraSelect Medical LLC.
ffimbres@ultraselectmed.com

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)