Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Sytec SRI CT Scanner

(6)

Brad Smith

Sytec SRI CT Scanner

August 20, 2010 09:42

Have a Sytec SRI CT Scanner. The gantry just keeps resetting as if someone hits the reset button on the TGP board. It's intermittent but when it starts there is absolutely no scanning and the system does not generate any errors. It will reset sometimes 10-15 times back to back. Anyone ever have this problem? Have replaced the TGP board but problem still exists. Help would be greatly appreciated.

Log inor Register

to rate and post a comment

(100)
(9)
(4)

JORGE LUIS LOPEZ

Re: Sytec SRI CT Scanner

August 20, 2010 03:30

is it just the ganry or the console as well ?

Log inor Register

to rate and post a comment

(6)

Brad Smith

Re: Sytec SRI CT Scanner

August 20, 2010 05:51

It is just the gantry, not the console. Thank you for your feedback.

Log inor Register

to rate and post a comment

Mario Silva

Re: Sytec SRI CT Scanner

August 20, 2010 08:40

Did you check the servo amp? some times it detects an overcurrent condition in the axial motor and stops or reset the gantry. Let me know what you find.

Log inor Register

to rate and post a comment

(6)

Brad Smith

Re: Sytec SRI CT Scanner

August 21, 2010 05:16

When I check everything, all is fine. I can sit there for an hour or two with no problems then all of the sudden it starts resetting. Still no errors and when I check voltages after it's reset everything is within spec. I did notice that the PDU has a fan not working so I've ordered another fan. Is it possible that a fan not working on the PDU (it's the small PDU, don't have the number on me)could be causing this intermittent problem?

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

Re: Sytec SRI CT Scanner

August 21, 2010 05:16

if the behavior is like press Reset button, you could track reset line of uP on the board, check noise on Vcc and check grounding.

it may be associated with rotation movement, put manual Sw on, and check while you change speed movements, or whie brake is actuated.

Log inor Register

to rate and post a comment

(6)

Brad Smith

Re: Sytec SRI CT Scanner

August 22, 2010 10:05

It will do it just while the system is just sitting idle. It also does it when scanning but will also do it when the system just sitting idle.

Log inor Register

to rate and post a comment

(72)
(1)

Elliot Marcus

Re: Sytec SRI CT Scanner

August 23, 2010 09:27

Sounds like a power supply. Can you watch the supply voltage when it's happening?

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

Re: Sytec SRI CT Scanner

August 25, 2010 09:57

ok, is power supply or connectors to board, because if you already exchanged TGP, then the problem is in the machine.

this kind of uP has a "Brown Out" circuit, when is activated behavior is the same like pushing reset button.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

We provide sales, service & repair for portable ultrasound systems & probes

If you have a broken Sonosite, Mindray, GE, Philips, Hitachi, or Aloka system or probe contact us today. Free ultrasound & probe EVALS with 1 Year Warranty on all repairs. We repair all orders in 1-5 days. Call 844-644-5440 Email: Sales@Lenard.Healthcare

DOTmed Certified stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (537 lots)