Vivid 3 Pro TX Tension Problem.

Tags: Cardiac Imaging

Mohamed Kocht

Vivid 3 Pro TX Tension Problem.

May 29, 2016 03:02

During the execution of the Vivid 3 Pro. an error message is displayed: Overload TX80 Voltage restarts the application. if the problem persists reboot. but when I Turn off the machine. it would not boot. (The central unit PC). the screen flashes "No Reports".
SEVERAL attempts after the machine started but after 10 minutes. the same problems. and she do not want to restart again. is that it is a supply problem of the central unit PC? Or other ?

Log inor Register

to rate and post a comment

(195)
(47)
(4)

EROTOKRITOS SKENTOS

re: Vivid 3 Pro TX Tension Problem.

May 30, 2016 10:36

The problem TX80 and TX40 is from the high voltage unit that is located in the board side underneath the low voltage power supply.

1 person liked this
Log inor Register

to rate and post a comment

Mohamed Kocht

re: re: Vivid 3 Pro TX Tension Problem.

May 31, 2016 10:24

Thanks for your answer, but how can fix it ? and same time the vivid work fine, with no problem.

Log inor Register

to rate and post a comment

(195)
(47)
(4)

EROTOKRITOS SKENTOS

re: re: re: Vivid 3 Pro TX Tension Problem.

June 01, 2016 10:31

The high voltage powersupply is for the probes,color and doppler.If one of the voltages fail,the 80 or 40,you will see symptoms like everythink works but i do not have color or everythink works but no doppler wave.If everythink works but you get this message it means that the high voltage is getting ready to fail permantly.

Log inor Register

to rate and post a comment

Mohamed Kocht

re: re: re: re: Vivid 3 Pro TX Tension Problem.

June 04, 2016 03:40

Thanks, But it's been 4 days I have not had a problem again. and vivid works.. what does it mean ?

Log inor Register

to rate and post a comment

(195)
(47)
(4)

EROTOKRITOS SKENTOS

re: re: re: re: re: Vivid 3 Pro TX Tension Problem.

June 06, 2016 10:09

That your high voltage powersupply is working now.This message can be produced from an bad inlet power or a probe that is consuming a lot of power or some bad components in the high voltage that might give you the same problem in the future.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

PHILIPS MX8000 & Brilliance Systems & Spare Parts In Stock Now!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com

DOTmed Certified stats
advertisement

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (537 lots)