Hologic Fluoroscan Insight FD C-arm Calibration bypass

March 13, 2023 10:19

Recently received a 2013 Hologic Fluoroscan Insight FD C-arm. System will not scan without calibration.

Is there a way to bypass the calibration without calling Hologic? I have the WIndows 7 disc, and the SW disc with me.

Any help is appreciated

Log inor Register

to rate and post a comment

(20)
(2)

xrayrepairdude

re: Hologic Fluoroscan Insight FD C-arm Calibration bypass

March 15, 2023 10:37

You should be able to calibrate the unit. Do you know if it is the x-ray controller that needs calibration or is it the detector that requires calibration? Either one can be done without too much trouble.

Ken

Log inor Register

to rate and post a comment

(7)
(2)
(2)

Brian Hoehn

re: re: Hologic Fluoroscan Insight FD C-arm Calibration bypass

March 15, 2023 11:59

I do not know which.

Log inor Register

to rate and post a comment

(20)
(2)

xrayrepairdude

re: Hologic Fluoroscan Insight FD C-arm Calibration bypass

March 16, 2023 10:17

I am based in central Florida, I can help if the machine is near there. Calibration can be a little finicky. Let me know if I can help.

Log inor Register

to rate and post a comment

(7)

(11)
(11)

Christopher Bacon

re: Hologic Fluoroscan Insight FD C-arm Calibration bypass

May 16, 2023 10:10

no you'll need to calibrate the unit before you can x-ray anything

Log inor Register

to rate and post a comment