How to connect GE Prodigy Compact DXA to PC

Tags: Bone Densitometer, GE

(12)

Darius

How to connect GE Prodigy Compact DXA to PC

August 11, 2025 11:33

Dear all,
I have a 2014 GE Prodigy Compact DEXA Scanner and I recenly had to upgrade the computer to a newer version. I am now running Encore version 16 on a Windows 7 Pro system, however, I am unable connect the scanner to the computer. The PC has a built in RS 232 port (COM3 port in windows). Everything is plugged in, I went through the installation steps in Encore, but but there is no communication between PC and DEXA scanner and the place where I can choose the com port is greyed out.
What am I missing? Do I need to make changes to the COM port configuration?
Any lead is greatly appreciated.
Thank you!
Darius

Log inor Register

to rate and post a comment

(4)
(2)

Ansh

re: How to connect GE Prodigy Compact DXA to PC

August 12, 2025 10:33

RS-232 Driver Missing or Incompatible

Even though Windows detects the COM port, Encore may require a specific serial driver (often provided by GE) for proper communication.

Check Device Manager → Ports (COM & LPT) and see if the COM3 port has a yellow exclamation mark.

If yes, install/update the correct serial port driver.

Wrong COM Port Number

Encore might be coded to only look at COM1 or COM2.

If your RS-232 is on COM3, you can change it:

Device Manager → Right-click COM3 → Properties → Port Settings → Advanced → Change COM Port Number to COM1 or COM2.

Restart PC and retry.

Encore COM Port Setup Locked

If the COM port setting in Encore is greyed out, it often means the scanner hardware driver/service didn’t load.

Check if the GE DEXA Service is running in Windows Services (services.msc).

Baud Rate / Serial Settings

RS-232 devices need matching settings. Try:

Baud rate: 115200 (or check GE manual)

Data bits: 8

Parity: None

Stop bits: 1

Flow control: None or Hardware

Log inor Register

to rate and post a comment

(12)

Darius

re: re: How to connect GE Prodigy Compact DXA to PC

August 28, 2025 11:00

Thank you so much. I checked windows services and was unable to find GE DEXA services. Also, when checking the COM-1 port in device manager. there is no yellow alert. I am unable to locate the specific GE COM driver on the Encore DVD, is there any other place where I can look?
Thank you again! Darius

Log inor Register

to rate and post a comment

(3)

(28)

A. Danisik

re: How to connect GE Prodigy Compact DXA to PC

August 12, 2025 10:34

Hi, we had the same issue on our system, please find the marks from our internal wiki below:

1. Verify COM port hardware and mapping
• In Windows Device Manager:
o Check under “Ports (COM & LPT)” if your RS-232 port shows up without warnings.
o Note the COM number Windows assigns (e.g., COM3).
o Right-click → Properties → Port Settings:
 Bits per second (baud rate): usually 115200 or 57600 for Prodigy
 Data bits: 8
 Parity: None
 Stop bits: 1
 Flow control: None
o If your built-in RS-232 shows as COM3 but Encore expects COM1, you can change it:
Properties → Port Settings → Advanced → Change COM Port Number → COM1 or COM2.

2. Encore software settings
• In Encore, the COM port selection is greyed out if:
o The scanner device driver (GE service driver) is not installed or not running.
o The software was installed in “network mode” instead of “local scanner” mode.
o Encore detects the port already in use or blocked by Windows.
• Solution:
o Re-run the Encore hardware setup from the installation media and select Local (Serial) Connection.
o Ensure you are logged in as Administrator when installing.
3. Driver requirement
• GE Prodigy scanners require a GE-specific serial communication driver that is separate from the generic Windows COM driver.
• This driver is normally installed automatically with Encore but may fail if:
o You upgraded Windows before installing Encore.
o You did not run the installer with Administrator rights.
• You can find the driver in the Encore install folder under something like:
C:\Program Files\GE Medical Systems\Encore\Drivers
and manually install it via Device Manager → Update Driver.

4. Hardware loopback test
To rule out cable or port failure:
• Use a simple RS-232 loopback plug (pin 2-3 shorted) and a terminal program (e.g., PuTTY) to confirm the PC’s COM port sends and receives data.
• If the loopback works, the issue is software or driver.
• If not, check the port in BIOS (many new boards have serial disabled by default).

5. BIOS settings
• On many newer PCs, the onboard RS-232 port is disabled in BIOS by default.
• Go into BIOS and ensure the serial port is enabled, and check which I/O address/IRQ it uses. Encore may only work with standard legacy COM assignments (e.g., COM1 = 3F8, IRQ 4).

Typical fix order for your case
1. Check BIOS → enable RS-232 → set to COM1 if possible.
2. In Windows Device Manager → set COM port to COM1 or COM2 → match baud and flow settings to GE defaults.
3. Reinstall Encore in local scanner mode, as Administrator, making sure the GE serial driver loads.
4. Test with loopback if no communication.


GE Prodigy COM Port Settings (Windows Device Manager → Port Settings)
• Bits per second (Baud rate): 115200
• Data bits: 8
• Parity: None
• Stop bits: 1
• Flow control: None

Advanced Port Settings:
• COM port number: COM1 or COM2 (Encore often defaults to COM1; it will not auto-detect if assigned to COM3+ without registry edit)
• FIFO buffers: Enabled (default), both transmit and receive set to maximum

BIOS Serial Port Settings (if applicable):
• Enabled
• I/O address for COM1: 3F8
• IRQ: 4

Log inor Register

to rate and post a comment

(12)

Darius

re: re: How to connect GE Prodigy Compact DXA to PC

August 28, 2025 11:01

Thank you so much for the detailed guidance. I've reviewed all the points, and everything is configured correctly except for the GE-specific communication driver.

I have renamed COM3 to COM1 in the Device Manager, and the port is currently using the generic Windows driver. However, I have been unable to locate the specific GE driver on the Encore DVD or the local machine.

Could you please advise on where else I might find this driver, or would it be possible to provide a copy?

Thank you for your assistance.

Best regards,
Darius

Log inor Register

to rate and post a comment

(3)

HD18

re: re: re: How to connect GE Prodigy Compact DXA to PC

August 28, 2025 12:06

I can check if, we have it. Do you see a Hardware ID, once you have connected the unit to the PC?

Log inor Register

to rate and post a comment

(12)

Darius

re: re: re: re: How to connect GE Prodigy Compact DXA to PC

August 29, 2025 09:52

No. When I connect the scanner to the PC, nothing happens, there is no entry in the device manager. thank you!

Log inor Register

to rate and post a comment

(12)

Darius

re: re: re: re: How to connect GE Prodigy Compact DXA to PC

August 29, 2025 09:52

Nothing happens when I connect the scanner to the PC, there is no entry in the device manager. Thank you!

Log inor Register

to rate and post a comment

(12)

Darius

re: re: re: re: re: How to connect GE Prodigy Compact DXA to PC

October 20, 2025 11:38

I have still not been able to make this work - anyone who can help with the ominous COM driver? Thanks!

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Brand new OEM Philips & Siemens Parts at unbeatable prices

Brand new OEM Philips & Siemens Parts at unbeatable prices shipping domestically and Includes warranty! Email P/N for pricing: sales@Relaxationinc.com (P) 813-925-1314, www.relaxationinc.com

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (533 lots)
Cbrx2sales
TIMED
Jun 17, 2026 (127 lots)