Over 500 Total Lots Up For Auction at One Location - NJ 06/15

voluson 730

Tags: Ultrasound

Posted by an Anonymous User

voluson 730

July 18, 2012 03:09

i have voluson 730 , bt02 , this system do not start up just writ on touch screen BOOTING , and on big screen try to start up and return from beginning i hope some body help me to start it to work

2 people liked this
Log inor Register

to rate and post a comment

RAFAEL SUAREZ

re: voluson 730

July 18, 2012 05:09

I have the service manual in PDF SEND ME YOUR EMAIL

1 person liked this
Log inor Register

to rate and post a comment

(12)

Khurram Siddiqui

re: re: voluson 730

March 02, 2013 09:09

Kindly send me this manual at khurramnisarsiddiqi@yahoo.com

Log inor Register

to rate and post a comment

(56)
(3)
(1)

morteza yousefi

re: re: voluson 730

October 11, 2013 11:53

this problem is for bep or main board.please check the sbc or main boRd.some times battery bord is defective.after it check the power cabel.

Log inor Register

to rate and post a comment

(12)

mauricio cid

re: re: voluson 730

March 25, 2014 08:01

Hi Friend, If you can send a copy of manual i appreciatte. Thanks... mauriciocidmx@hotmail.com

Log inor Register

to rate and post a comment

(1)

Pam Tulley

re: re: voluson 730

August 03, 2018 10:12

Rafael, I am trying to service my bios battery and am getting no help from the tech support, since they want to get paid to do it for me. Can you send me your service manual PDF so I can quit messing with them. rtulley@hotmail.com. thank you in advance. Pam Tulley

Log inor Register

to rate and post a comment

Ricardo A. Sanchez

re: voluson 730

July 18, 2012 05:45

probably the BIOS/CMOS Battery ran out, you need to reaplce and load the right settigns on the BIOS, I have seen these kind of problems before.

Log inor Register

to rate and post a comment

(1)

Ashvin Contractor

re: re: voluson 730

October 04, 2015 09:58

Some how the SBC board in our Voluson 730 (4D) does not have any battery. I looked at the youtube video as well the GE Service manual and the SBC Board shown in them are different than I have. Our machine was bought in 2001.
I would like to attach a photo of the SBC Board to this email but cannot find any option to do so.
Any help would be appreciated from any of the service engineers.

Log inor Register

to rate and post a comment

Mario Wug

re: voluson 730

July 19, 2012 12:18

I would suggest to reload software. Just remember the BT02 systems use a boot CD to start the load procedure.

Log inor Register

to rate and post a comment

(59)
(5)
(4)

Abdelmuin Khatib

re: re: voluson 730

July 19, 2012 07:58

hello
thank you for your help , i have CD for software , but how i can reload it ?

Log inor Register

to rate and post a comment

Mario Wug

re: re: re: voluson 730

July 19, 2012 04:14

First do the BIOS check thats really well explain there. If that doesn't work then the SW load will be the next step. Let me see if I have the loading procedure.

Log inor Register

to rate and post a comment

(37)
(4)

carlos mendez

re: re: re: voluson 730

July 17, 2013 09:15

https://www.youtube.com/watch?v=VvmLiN1zj3Y&list=PL5196474BA82113E3&index=1

here is video how to reload software, when it boot you have to enter the bios pass its Rudi, you have to select the system to boot with an optical device.

Log inor Register

to rate and post a comment

(24)
(3)

Joseph Lovett

re: voluson 730

July 19, 2012 08:00

I have encountered a similar issue and Mr. Sanchez is likely correct. It should be easy to check however. As the system begins to boot you need to enter the BIOS settings. The system does not tell you what key to press but it is either "F1", "F2", or "Del". The best thing to do is to press each one in sequence repeatedly until the BIOS screen is displayed. Once this comes up, choose the option to "load defaults". Do this and then choose the option to "save settings and exit". The system should then reboot normally.

If this works, then the problem is definitely the BIOS battery. It is a CR2032 coin battery that is located on the SBC (Single Board Computer) inside the system cabinet. Opening the system cabinet will require opening the back of the system, disconnecting all of the cables attached to it and pulling the entire box out. Voluson did not make it easy.

Good luck with this.

Log inor Register

to rate and post a comment

(59)
(5)
(4)

Abdelmuin Khatib

re: re: voluson 730

July 19, 2012 12:52

really thank you , i will do what you inform me.

Log inor Register

to rate and post a comment

(59)
(5)
(4)

Abdelmuin Khatib

re: re: re: voluson 730

July 20, 2012 08:45

i do what you said and press many times F1 , F2 , DELL but system do not go to bios , but if we press ctrl&a , system go to SCSI .
when system start up write on screen SCSI BIOS NOT INSTALLED

Log inor Register

to rate and post a comment

(24)
(3)

Joseph Lovett

re: re: re: re: voluson 730

July 20, 2012 11:38

The "SCSI BIOS NOT INSTALLED" message is of no matter. Are you pressing the F1, F2, Del as soon as you turn the system on? Each button must be pressed separately.

Log inor Register

to rate and post a comment

Ricardo A. Sanchez

re: re: re: re: voluson 730

July 20, 2012 03:42

as sson as the system starts press the Key DEL, that will take you to the BIOS, yit may ask you for a password, Rudi

Log inor Register

to rate and post a comment

(9)

Zenon Coronel

re: re: voluson 730

November 30, 2014 01:02

I have a voluson 730 expert BT01 the system does boot ok, but touch panel not working, only work keyboard and mouse
What can I do?
Thanks

Log inor Register

to rate and post a comment

(56)
(3)
(1)

morteza yousefi

re: voluson 730

July 20, 2012 08:42

Dear
you must chek the main pord of the pc or sbc bord.
it is error for main bord or sbc bord

Log inor Register

to rate and post a comment

Mario Wug

re: re: voluson 730

July 20, 2012 03:11

Yeah. You can try to take SBC out and clean it well. Reseat it and try again. Sometimes those boards get bad connections.

Log inor Register

to rate and post a comment

(59)
(5)
(4)

Abdelmuin Khatib

re: re: re: voluson 730

July 24, 2012 09:57

hi
i enter to BIOS but i do not know what i must do , any way this system try to start but do not go on and blue screen appear written (beginning dump of physical memory
physical memory a )
so are there problem in memory?

Log inor Register

to rate and post a comment

Mario Wug

re: re: re: re: voluson 730

July 24, 2012 12:24

When on BIOS, you have to load optimal configuration and save and exit. After that the system might boot without problem.

Log inor Register

to rate and post a comment

(24)
(3)

Joseph Lovett

re: re: re: re: voluson 730

July 24, 2012 12:24

When you are in the BIOS, there will be a selection to load the default setup. Choose that. There will then be a selection to save the settings and exit. Choose that. It will probably ask for you to confirm this (Y/N). Type "Y". The system should then reboot and hopefully come up normally. If it does, then you have only fixed the problem temporarily. The BIOS battery on the SBC will need to be replaced.

Log inor Register

to rate and post a comment

(59)
(5)
(4)

Abdelmuin Khatib

re: re: re: re: re: voluson 730

July 24, 2012 12:41

thank you for your help, i did that but still same problem the system do not go on start to go but return start again

Log inor Register

to rate and post a comment

(59)
(5)
(4)

Abdelmuin Khatib

re: re: re: re: re: voluson 730

July 24, 2012 12:41

hello
i did but the problem still same and system do not go on

Log inor Register

to rate and post a comment

(24)
(3)

Joseph Lovett

re: re: re: re: re: re: voluson 730

July 24, 2012 05:31

I think that it is possible then that either the operating system is corrupt of the system hard drive has failed.

Log inor Register

to rate and post a comment

(59)
(5)
(4)

Abdelmuin Khatib

re: re: re: re: re: re: re: voluson 730

July 25, 2012 10:02

OH, CAN WE CHANGE IT EASY AND HOW WE CAN FIND IT?

Log inor Register

to rate and post a comment

(24)
(3)

Joseph Lovett

re: re: re: re: re: re: re: re: voluson 730

July 25, 2012 12:50

If reloading software is possible, I would try that first. Replacing anything on this system is a challenge. Everything is contained within the large stainless steel box within the base of the unit. All of the wires and cables must be disconnected from this box (mark which ones plug into where) and it must be removed from the unit before it can be opened. Make sure that all of the probes have been removed. All of the system components including the hard drive are contained within.

Log inor Register

to rate and post a comment

Mario Wug

re: re: re: re: re: re: voluson 730

July 25, 2012 10:14

Reload SW would be your best option now.

Log inor Register

to rate and post a comment

(59)
(5)
(4)

Abdelmuin Khatib

re: re: re: re: re: re: re: voluson 730

July 25, 2012 12:50

i have two CD for software but how we can reload it i put cd in cd reader but no think happen

Log inor Register

to rate and post a comment

Mario Wug

re: re: re: re: re: re: re: re: voluson 730

July 25, 2012 02:56

You have to change the BIOS setting to boot with the CD.

Log inor Register

to rate and post a comment

Mario Wug

re: re: re: re: re: re: re: voluson 730

July 25, 2012 12:50

One important thing you need to have when reloading SW on the system, is to have all your options listed with their keys.

1 person liked this
Log inor Register

to rate and post a comment

(9)

Zenon Coronel

re: re: re: re: re: re: re: re: voluson 730

March 01, 2013 02:17

where is the key

Log inor Register

to rate and post a comment

Mario Wug

re: re: re: re: re: re: re: re: re: voluson 730

March 01, 2013 05:24

You have to go to settings, and then system. You should find options listed there. Print them or copy them.

Log inor Register

to rate and post a comment

(9)

Zenon Coronel

re: re: re: re: re: re: re: re: re: re: voluson 730

March 22, 2013 06:00

how I enter to setting and then system
I have voluson 730 expert BT01
I have software cd but no boot cd

Log inor Register

to rate and post a comment

(9)

Zenon Coronel

re: re: re: re: re: re: re: re: re: re: re: voluson 730

March 23, 2013 06:44

I have voluson 730 BT01
the system do not go on start to go but return start again
The BIOS battery on the SBC is new

Log inor Register

to rate and post a comment

(56)
(3)
(1)

morteza yousefi

re: voluson 730

November 01, 2013 05:24

frend.i check your text.do you want work your system or no.i see some body say some ting that all is not curect.check step by step.1.open the back door.2.check the 6 lamp of voltag.if 6 of them is ok.if not ok and 1 or some of them be off check the pwer supplay.if is ok open the Bep and check the voltag by voltag cheker for the moder bord.if all voltag is not ok the cpp bord is de fectiv.if you have all reng of voltag and evry thing is ok the main bord i100% is defective.cheng it.
yousefi

Log inor Register

to rate and post a comment

(9)

Zenon Coronel

re: voluson 730

November 30, 2014 01:02

I have a voluson 730 expert BT01, the system does boot ok but touch panel not working
What can I do?
Thanks

Log inor Register

to rate and post a comment

Mitesh Chandarana

re: voluson 730

February 24, 2017 10:36

GE Voluson 730 Expert - "OCP Connection Failed" Error shows up after five attempts of "Connecting to Console"
Can someone help finding solution on this? Thanks

Log inor Register

to rate and post a comment

Ricardo A. Sanchez

re: re: voluson 730

February 26, 2017 02:35

you have connected something wrong, the control panel can't connect to the GET box

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Probo Medical | We Buy, Repair, and Sell New and Used Ultrasound Equipment

We provide sales, service, rental, and repair for ultrasound systems, probes, and C-arms. We have the largest inventory in the world. Contact us about our special offers on certified refurbished equipment! Call: 317-759-9210 Email: sales@probomedical.com.

stats
advertisement

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)