My Eppendorf centrifuge 5804R showing no rotor and error 3 when start is pressed. The rotor dont move at all but can hear the compressor sound when press start. Please help.
Here is what the service manual says about error 3
Error 3
no rotor
No rotor pulses during run or
acceleration.
Missing rotor, measuring system for
rotational speed defective, drive
defective.
– Rotor not fastened.
– No pulses after start. Device brakes
for 15 s.
– Pulse failure during run. Free deceleration
requires 12 min!
Error remains active after a power
supply interruption.
– Fasten rotor.
Keep device switched on and start
again after > 12 min.
– Service program 8.
– Check/replace motor cap with speed
sensor, check connections.
– Replace "Control/FC" PCB,
check connector X10 on PCB
"Control".
– Check run of motor, replace motor if
necessary.
no rotor
No rotor pulses during run or
acceleration.
Missing rotor, measuring system for
rotational speed defective, drive
defective.
– Rotor not fastened.
– No pulses after start. Device brakes
for 15 s.
– Pulse failure during run. Free deceleration
requires 12 min!
Error remains active after a power
supply interruption.
Remedy
– Fasten rotor.
Keep device switched on and start
again after > 12 min.
– Service program 8.
– Check/replace motor cap with speed
sensor, check connections.
– Replace "Control/FC" PCB,
check connector X10 on PCB
"Control".
– Check run of motor, replace motor if
necessary.
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Bubacarr Sissoko
centrifuge 5804r showing error 3
August 20, 2014 08:25
My Eppendorf centrifuge 5804R showing no rotor and error 3 when start is pressed. The rotor dont move at all but can hear the compressor sound when press start. Please help.
to rate and post a comment
Jacob Reinhart
re: centrifuge 5804r showing error 3
August 21, 2014 12:51
Here is what the service manual says about error 3
Error 3
no rotor
No rotor pulses during run or
acceleration.
Missing rotor, measuring system for
rotational speed defective, drive
defective.
– Rotor not fastened.
– No pulses after start. Device brakes
for 15 s.
– Pulse failure during run. Free deceleration
requires 12 min!
Error remains active after a power
supply interruption.
– Fasten rotor.
Keep device switched on and start
again after > 12 min.
– Service program 8.
– Check/replace motor cap with speed
sensor, check connections.
– Replace "Control/FC" PCB,
check connector X10 on PCB
"Control".
– Check run of motor, replace motor if
necessary.
to rate and post a comment
Long Nguyen Kim
re: re: centrifuge 5804r showing error 3
September 07, 2016 12:13
My Eppendorf centrifuge 5415R showing Error 3 and can not control anything when I turned the switched on.
Please help me fix this error! Thanks!
to rate and post a comment
Jacob Reinhart
re: centrifuge 5804r showing error 3
August 21, 2014 12:51
Error 3
Cause
no rotor
No rotor pulses during run or
acceleration.
Missing rotor, measuring system for
rotational speed defective, drive
defective.
– Rotor not fastened.
– No pulses after start. Device brakes
for 15 s.
– Pulse failure during run. Free deceleration
requires 12 min!
Error remains active after a power
supply interruption.
Remedy
– Fasten rotor.
Keep device switched on and start
again after > 12 min.
– Service program 8.
– Check/replace motor cap with speed
sensor, check connections.
– Replace "Control/FC" PCB,
check connector X10 on PCB
"Control".
– Check run of motor, replace motor if
necessary.
Let me know if we can help.