Hello,
Someone have an experience on Siemens Iconos R100, I have a code error 240/032.
The table not move no more, but all voltajes on board basic unit is working,
I try to connect with service but message on pc that "not all parts working"
thanks
all safety breakere stick is good, all voltajes is present but the error is on table,
only I found the voltaje on U20, U7 is 210VAC not 240.
does this is the normal?
Reason, Cause:
- Emergency stop switch pressed.
- Safety breaker has responded (unit has moved into the safety circuit).
- Limit switch defective or a cable break
- 24 V~ for the safety breaker missing
Action:
- 24 V~ for the safety breaker missing: test points A4.D31.XP14 versus XP2; F10 fuse on the A4.T1 transformer OK? (thermoswitches T1.S1 to T1.S3 OK?)
- Move the unit out of the safety circuit.
- Check limit switches and cabling.
i check all system but on D1 B10 red light on and 240/032 show me.
some times the system turn on good no error, but many times the control appear error 240/032
GCI offers complete MRI & CT sales & service with ROI analysis, in-house financing, contractor drawings, accreditation, training, marketing, and OEM-level support beyond the sale.
Germni
Error 240/032 Siemens Iconos R100
September 24, 2018 08:18
Hello,
Someone have an experience on Siemens Iconos R100, I have a code error 240/032.
The table not move no more, but all voltajes on board basic unit is working,
I try to connect with service but message on pc that "not all parts working"
thanks
to rate and post a comment
Bill Masters
re: Error 240/032 Siemens Iconos R100
September 24, 2018 08:47
rror 240/032 (Iconos R100)
Meaning:
Basic unit: Safety breaker sticks when system is switched on.
Reason, Cause:
Unit is not operational. Safety breaker sticks.
Action:
- Switch the system off and back on.
- Check the safety breaker; if necessary, replace it.
to rate and post a comment
Germni
re: re: Error 240/032 Siemens Iconos R100
September 25, 2018 09:34
all safety breakere stick is good, all voltajes is present but the error is on table,
only I found the voltaje on U20, U7 is 210VAC not 240.
does this is the normal?
to rate and post a comment
Germni
re: re: Error 240/032 Siemens Iconos R100
September 26, 2018 09:51
Does the safety breaker sticks is K1 on basic unit?
or K1 on A1 D2 on SPOR FILM DEVICE
to rate and post a comment
Bill Masters
re: Error 240/032 Siemens Iconos R100
September 24, 2018 08:47
Error 241/032
Meaning:
Basic unit: safety breaker deenegized.
Reason, Cause:
- Emergency stop switch pressed.
- Safety breaker has responded (unit has moved into the safety circuit).
- Limit switch defective or a cable break
- 24 V~ for the safety breaker missing
Action:
- 24 V~ for the safety breaker missing: test points A4.D31.XP14 versus XP2; F10 fuse on the A4.T1 transformer OK? (thermoswitches T1.S1 to T1.S3 OK?)
- Move the unit out of the safety circuit.
- Check limit switches and cabling.
to rate and post a comment
Germni
re: re: Error 240/032 Siemens Iconos R100
September 26, 2018 09:37
i check all system but on D1 B10 red light on and 240/032 show me.
some times the system turn on good no error, but many times the control appear error 240/032
to rate and post a comment