Error 240/032 Siemens Iconos R100

September 24, 2018 08:18

Hello,
Someone have an experience on Siemens Iconos R100, I have a code error 240/032.
The table not move no more, but all voltajes on board basic unit is working,
I try to connect with service but message on pc that "not all parts working"
thanks

Log inor Register

to rate and post a comment

Bill Masters

re: Error 240/032 Siemens Iconos R100

September 24, 2018 08:47

rror 240/032 (Iconos R100)

Meaning:
Basic unit: Safety breaker sticks when system is switched on.

Reason, Cause:
Unit is not operational. Safety breaker sticks.

Action:
- Switch the system off and back on.
- Check the safety breaker; if necessary, replace it.

Log inor Register

to rate and post a comment

(148)
(25)

Germni

re: re: Error 240/032 Siemens Iconos R100

September 25, 2018 09:34

all safety breakere stick is good, all voltajes is present but the error is on table,
only I found the voltaje on U20, U7 is 210VAC not 240.
does this is the normal?

Log inor Register

to rate and post a comment

(148)
(25)

Germni

re: re: Error 240/032 Siemens Iconos R100

September 26, 2018 09:51

Does the safety breaker sticks is K1 on basic unit?
or K1 on A1 D2 on SPOR FILM DEVICE

Log inor Register

to rate and post a comment

Bill Masters

re: Error 240/032 Siemens Iconos R100

September 24, 2018 08:47


Error 241/032

Meaning:
Basic unit: safety breaker deenegized.

Reason, Cause:
- Emergency stop switch pressed.
- Safety breaker has responded (unit has moved into the safety circuit).
- Limit switch defective or a cable break
- 24 V~ for the safety breaker missing

Action:
- 24 V~ for the safety breaker missing: test points A4.D31.XP14 versus XP2; F10 fuse on the A4.T1 transformer OK? (thermoswitches T1.S1 to T1.S3 OK?)
- Move the unit out of the safety circuit.
- Check limit switches and cabling.

Log inor Register

to rate and post a comment

(148)
(25)

Germni

re: re: Error 240/032 Siemens Iconos R100

September 26, 2018 09:37

i check all system but on D1 B10 red light on and 240/032 show me.
some times the system turn on good no error, but many times the control appear error 240/032