Dear all ,
Please i have a probleme with a panoramic dental Gendex ; i cannot finish a cycle for exposure =>the tube and the receptor stopped and it give me error ;Panoramic stand xray break .
please if you have any idea for the this problem reply me.
also if you have the service manuel for this machine contact me .
THANKS
Maybe the problem is one of the micro switch that controls the movement. They are located under the top cover. You can easly remove tha cover, 4 allen screws. Place the machine in the start positions, then order a turn without Xrays and check were it stops and the correct switch action.
thanks Mr fernando for you answers , i will check the micro switch i would like to told you that it does not stopped in a single position ,but it stopped in different position sometimes at the beginning and sometimes in the middle.
also do you can send me the service manuel(if you have it )by mail ; daly2251@yahoo.fr
Hi Fernado, can you pls email me with the service manual, I'm having image quality issue with my unit, I highly appreciate if you could do that. thanks.
I have the cr, the problem is that sometimes the images are good and sometimes it's kind a white wash. I was thinking to check the kV & mA regulation to see if its stable. I can't do it without manual. Highly appreciated your help.
We have some problem in gendex orthoralix sd2 when we turn ON it after a few seconds the following error appears on control panel
SYSTEM CPU FAIL
E2PROM CHECKSUM
re: re: Re: Gendex orthoralix error & service manual
May 02, 2012 05:58
This mean that you have a problem with E2PROM on the microprocessor board. You can reset or replace it. If this didn't help you must replace microprocessor board.
If you replace or reset the E2PROM you must recalibrate the system - tube calibration, all movements ....
re: re: re: Re: Gendex orthoralix error & service manual
May 03, 2012 10:02
With reset no progress same error can you please send me some direction where is it located (E2PROM) and after replacement of E2PROM what process or procedures should i have to done
re: re: re: Re: Gendex orthoralix error & service manual
June 13, 2016 02:12
Hi Niki, have you service manual?... can you please email me with manual?... I have problem with. I don´t have the hand control and I have to make it again and I don´t know the connections on the boards. Thanks
re: re: Re: Gendex orthoralix error & service manual
August 31, 2018 04:32
Hi - I also have a problem with image quality. Image is fuzzy, can't seem to get a clear enough image for diagnostic quality. Imaging software can only do so much when the source image is not sharp enough. Gendex helped to callibrate. This is a used pano and they don't make all the parts anymore. What part affects sharpness? Any ideas? Maybe someone can help all of us dealing with this.
8500 DDE
Latest Gendex suggested I order a cable that connects the PA7 board to the PA6 board. I hope that helps. This is a used pano but had low usage. Does this sound like it would work? And is there anything else I could try?
Specializing in GE, we keep your ICU monitors calibrated and patient-ready—fast, reliable service from expert biomed techs. Southeastern Biomedical: 888-310-7322.
redissi daly
Gendex orthoralix error & service manual
April 19, 2009 06:13
Dear all ,
Please i have a probleme with a panoramic dental Gendex ; i cannot finish a cycle for exposure =>the tube and the receptor stopped and it give me error ;Panoramic stand xray break .
please if you have any idea for the this problem reply me.
also if you have the service manuel for this machine contact me .
THANKS
to rate and post a comment
Fernando Canga
Re: Gendex orthoralix error & service manual
April 23, 2009 09:20
Hello
Maybe the problem is one of the micro switch that controls the movement. They are located under the top cover. You can easly remove tha cover, 4 allen screws. Place the machine in the start positions, then order a turn without Xrays and check were it stops and the correct switch action.
I have the manual in case you need it.
to rate and post a comment
redissi daly
Re: Gendex orthoralix error & service manual
April 23, 2009 04:18
thanks Mr fernando for you answers , i will check the micro switch i would like to told you that it does not stopped in a single position ,but it stopped in different position sometimes at the beginning and sometimes in the middle.
also do you can send me the service manuel(if you have it )by mail ; daly2251@yahoo.fr
Thanks
to rate and post a comment
Fernando Canga
Re: Gendex orthoralix error & service manual
April 24, 2009 08:46
Hi,
Did you check the continuity of the X-ray hand swicht cable?
I don´t have the manual in *.doc, neither *.pdf format. But in case you still need it I could scan all the manual, let me know in that case...
Regards
to rate and post a comment
montano aguirre
Re: Gendex orthoralix error & service manual
March 30, 2012 06:13
Hi Fernado, can you pls email me with the service manual, I'm having image quality issue with my unit, I highly appreciate if you could do that. thanks.
to rate and post a comment
Fernando Canga
Re: Gendex orthoralix error & service manual
April 02, 2012 08:56
Hi,
Maybe you have a problem with the movement or the speed. Do you have CR or conventional film?
I´ll try to scan the manuel but it will take time ´cause there are a lot of pages.
Regards
to rate and post a comment
montano aguirre
Re: Gendex orthoralix error & service manual
April 02, 2012 05:39
I have the cr, the problem is that sometimes the images are good and sometimes it's kind a white wash. I was thinking to check the kV & mA regulation to see if its stable. I can't do it without manual. Highly appreciated your help.
Thanks,
Montano
to rate and post a comment
TOM CHARLES
re: Re: Gendex orthoralix error & service manual
May 02, 2012 09:22
We have some problem in gendex orthoralix sd2 when we turn ON it after a few seconds the following error appears on control panel
SYSTEM CPU FAIL
E2PROM CHECKSUM
what should be the solution for this
to rate and post a comment
Alexander Tonev
re: re: Re: Gendex orthoralix error & service manual
May 02, 2012 05:58
This mean that you have a problem with E2PROM on the microprocessor board. You can reset or replace it. If this didn't help you must replace microprocessor board.
If you replace or reset the E2PROM you must recalibrate the system - tube calibration, all movements ....
to rate and post a comment
TOM CHARLES
re: re: re: Re: Gendex orthoralix error & service manual
May 03, 2012 10:02
With reset no progress same error can you please send me some direction where is it located (E2PROM) and after replacement of E2PROM what process or procedures should i have to done
thanks
greetings
to rate and post a comment
oktay ayhan
re: Re: Gendex orthoralix error & service manual
March 12, 2016 12:17
hello, how are you need me Gendex orthoralix 9200 service manual. Can you help me?
to rate and post a comment
Niki
re: re: Re: Gendex orthoralix error & service manual
March 13, 2016 07:23
Your e-mail?
to rate and post a comment
Marcos Giorgio
re: re: re: Re: Gendex orthoralix error & service manual
June 13, 2016 02:12
Hi Niki, have you service manual?... can you please email me with manual?... I have problem with. I don´t have the hand control and I have to make it again and I don´t know the connections on the boards. Thanks
to rate and post a comment
Marcos Giorgio
re: re: re: Re: Gendex orthoralix error & service manual
June 13, 2016 06:10
Dear Niki, my e-mail is medtech.bioingenieria@gmail.com
Thanks!!!
to rate and post a comment
Nadeem Hameed
re: re: re: Re: Gendex orthoralix error & service manual
December 07, 2018 08:55
hi Niki ,please snd me error list and service manual of gendex orthoralix 9000
to rate and post a comment
Omar Laoudini
re: Re: Gendex orthoralix error & service manual
March 21, 2016 03:46
Hi Mr Fernando,
I have a problem with image quality of gendex orthoralix sd2, i think the resolution.
Please help me.
to rate and post a comment
Karim Chebli
re: re: Re: Gendex orthoralix error & service manual
April 17, 2018 05:59
Hi I have also a problem a distorted image can any ne could help me with that?
to rate and post a comment
Susan Aragaki
re: re: Re: Gendex orthoralix error & service manual
August 31, 2018 04:32
Hi - I also have a problem with image quality. Image is fuzzy, can't seem to get a clear enough image for diagnostic quality. Imaging software can only do so much when the source image is not sharp enough. Gendex helped to callibrate. This is a used pano and they don't make all the parts anymore. What part affects sharpness? Any ideas? Maybe someone can help all of us dealing with this.
to rate and post a comment
Niki
re: re: re: Re: Gendex orthoralix error & service manual
September 07, 2018 10:49
First whats your Orthoralix model?
Do you have a calibration phantom?
to rate and post a comment
Susan Aragaki
re: re: re: re: Re: Gendex orthoralix error & service manual
September 08, 2018 03:05
8500 DDE
Latest Gendex suggested I order a cable that connects the PA7 board to the PA6 board. I hope that helps. This is a used pano but had low usage. Does this sound like it would work? And is there anything else I could try?
to rate and post a comment
Marcos Giorgio
re: Re: Gendex orthoralix error & service manual
June 13, 2016 02:11
Hola Fernando. I need a service manual to know how is a connection the hand control. I don´t have the hand control and I have to make it again. Thanks
to rate and post a comment
Marcos Giorgio
re: Gendex orthoralix error & service manual
June 13, 2016 02:11
Hello to all. Anyone knows how is the conection of de hand control?.. and anyone got the service manual?.. Thanks to all
to rate and post a comment
joe ssali
re: re: Gendex orthoralix error & service manual
June 14, 2016 05:03
I have the service manual but in hard copy. I can scan it for you.
to rate and post a comment
diego estrada
re: Gendex orthoralix error & service manual
February 15, 2019 09:09
anyone knows what is the following error
Fault detected # 1 ?
to rate and post a comment