Help resolve Error Message Appelem Carmarque X-ray

Tags: Radiology - R/F, APELEM

(7)
(1)

AAP

Help resolve Error Message Appelem Carmarque X-ray

December 15, 2022 11:40

Hi,
Our Appelem Carmarque X-ray console displays " PS contact error' .
It has a Magnum C/ Magnum C DR series x-ray generator.

Many thanks.

Log inor Register

to rate and post a comment

(5)
(3)

Pascal

re: Help resolve Error Message Appelem Carmarque X-ray

December 16, 2022 11:00

Dear AAP
How are you?
Is the message displayed during the initialisation system?
Best regards

2 people liked this
Log inor Register

to rate and post a comment

(7)
(1)

AAP

re: re: Help resolve Error Message Appelem Carmarque X-ray

December 21, 2022 10:12

Pascal,
Sorry I'm just seeing your chat.
Yes it is during the initialisation .

Log inor Register

to rate and post a comment

(62)
(14)

MM

re: Help resolve Error Message Appelem Carmarque X-ray

December 19, 2022 11:04

Main contactor does not switch on
First Check all power suplies

1 person liked this
Log inor Register

to rate and post a comment

(7)
(1)

AAP

re: re: Help resolve Error Message Appelem Carmarque X-ray

December 21, 2022 10:12

Ok thanks. How do we switch it on?

Log inor Register

to rate and post a comment

(7)
(1)

AAP

re: re: Help resolve Error Message Appelem Carmarque X-ray

January 23, 2023 10:19

Hi,
We located CONTACTOR supplied by F1,F2 ,F3 through an inductor before getting into the contactor that we can see and there is voltage coming to this but there is no voltage coming from JD10(1) and JD10 (3)to the coil of the main contactor.
1. Is this an exposure contactor or mani power contactor.
2. How do we switch it on.
3. Kindly advice on what to do.

Sorry, the system is installed in a remote mission Hospital in Africa with no access to qualified technical assistance.
Very many thanks.

Log inor Register

to rate and post a comment

X-Ray Tube Replacements for GE, Siemens, Varian, Philips, Toshiba, Shimadzu

Over thirty five years in the X-Ray tube manufacturing. X-Ray tube replacements for all leading imaging equipment manufactures. From the past twenty years to today's modern mobile digital imaging equipment. Overnight shipping. SC, USA. 843 767 8090

advertisement

Upcoming Auction Events

Tekyard Medical
TIMED Ending Soon
Feb 08, 2023 (220 lots)
Global Medical Devices LLC
TIMED
Feb 10, 2023 (83 lots)
SDI Medical Consultants
TIMED
Feb 13, 2023 (116 lots)
Unitedrecovery.ca
TIMED
Feb 14, 2023 (72 lots)
Therapeutic Technologies
TIMED
Feb 17, 2023 (2053 lots)