Hologic QDR 1000 Plus

(18)
(2)
(2)

Ray Gonzalez

Hologic QDR 1000 Plus

October 05, 2004 08:35

Need help with a Hologic QDR 1000 Plus. I am getting the error ("Timeout: no 120Hz interrupts. This points out to the Timing and Control PCB.I have replaced all the boards on the unit including the computer and still get the same error.
Any help will be apreciatted.
Thanks in advance.
Please reply to: Ray Gonzalez
ray@4biomed.com

Log inor Register

to rate and post a comment

(128)
(10)
(5)

Manuel Esquer

Re: Hologic QDR 1000 Plus

October 13, 2004 02:32

Verify chopper pickoff sincronization
Check filter wheel pickoff
Check timing & control board and reseat

Log inor Register

to rate and post a comment

(18)
(2)
(2)

Ray Gonzalez

Re: Hologic QDR 1000 Plus

October 13, 2004 07:05

We appreciate your help Manuel. The problem was the x-ray tank. We just replaced it thanks to the fact that we have one unit for parts.

Regards,
Ray Gonzalez

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: Cbrx2sales Surplus Auctions July 15th

Cbrx2sales closeout auctions take place on DOTmed.com. Bid today on 125 lots of medical equipment. Onsite or virtual inspection available.

stats

Upcoming Auction Events

Medical Equipment Superstore
TIMED
Jul 14, 2026 (501 lots)
Cbrx2sales
TIMED
Jul 15, 2026 (124 lots)
PBC Medical Solutions
TIMED
Jul 17, 2026 (11 lots)
C2 Management
TIMED
Jul 21, 2026 (345 lots)
Surplus Service
TIMED
Jul 22, 2026 (136 lots)