SEARCH
Current Location:
>
>
> OEC9900

OEC9900

(44)
(6)

Francesco Didonato

OEC9900

May 13, 2019 10:39

Hello i have problem with oec 9900 at boot touch-screen don't working when i touch it and of tra to fluoro after fluoro COMUNICATION error somebody can help me about this problem?
Thank's

Log inor Register

to rate and post a comment

(9)

ALLEN GODFREY

re: OEC9900

May 15, 2019 10:25

I would start by checking the usb cable and connections from the touch screen monitor to the GPOS CPU. GPOS is the cpu with the usb extender on it. the touch screen should be one of the cables plugged into the cpu onboard usb jacks. I might be more helpfull if I have more details after you try this. arg4team8@bellsouth.net

Log inor Register

to rate and post a comment

General Biomedical Service - New & Completely Truly Refurbished Ventilators

GBS offers service, repair, rentals & spare parts for your ventilators. With over 25 years experience our certified & factory-trained biomedical technicians refurbish all our used ventilators. Call today 800.558.9449

 

advertisement

 

 

advertisement

 

advertisement
advertisement
Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2019 DOTmed.com, Inc.
ALL RIGHTS RESERVED