Ronald O'Leary

Philips Super 80cp X-Ray Generator

March 16, 2023 10:17

Does anyone have a service manual with error codes for a Philips Super 80CP X-Ray generator that they could share?
Thanks!

Log inor Register

to rate and post a comment

(68)
(12)

Herrmann

re: Philips Super 80cp X-Ray Generator

March 17, 2023 10:18

Hi,
What is the problem by your SCP?

Log inor Register

to rate and post a comment

(5)

Ronald O'Leary

re: re: Philips Super 80cp X-Ray Generator

March 20, 2023 10:42

Generator error 3910, error cleared with power cycle. Service manual would be of great assistance. Thanks

Log inor Register

to rate and post a comment

(68)
(12)

Herrmann

re: re: re: Philips Super 80cp X-Ray Generator

March 20, 2023 04:07

Hi Ronald,
Now, this is a Error that was often in the field.
Look to the frequency relay in the lower part of
the converter under the capaciter battery.
And mostly was the Relais K7 on the EZ112 defect.
Is there a electronic relais ? When no, then in the
past was aviable by Philips a update kit.
I wish you success

Log inor Register

to rate and post a comment

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs.

Medilab Global provides high-quality pre-owned imaging equipment, parts, and service to the healthcare industry worldwide. Our number 1 position is to provide excellent customer service.. 305-234-0084 | Email: info@medilabglobal.com | Hablamos Español

DOTmed Certified DOTmed 100
advertisement

Upcoming Auction Events

Surplus Service
TIMED Ending Soon
Mar 27, 2023 (71 lots)
Tekyard Medical
TIMED
Mar 29, 2023 (214 lots)
Therapeutic Technologies
TIMED
Mar 30, 2023 (1553 lots)
ZRG Medical
TIMED
Apr 07, 2023 (78 lots)