Siemens Somatom Perceptive 128 slice Main Breaker trapping

Tags: CT, SIEMENS

(1)

Khan Biomed

Siemens Somatom Perceptive 128 slice Main Breaker trapping

June 07, 2021 12:27

Hello Guys,
The problem happened during sudden power loss .
The Somatom Perspective 128 slice main breaker is trapping during Syn.
We replaced the main breaker but didn't work and the same issue again.
The nature of the problem is :
When we power ON the Gantry and it starts Syn with PC and the moment the K3 starts to energize ,The main breaker traps and K2 is not activating either.
I Check the PDS BUT didn't find any physical damage on it ,so i guess the issue with PDS itself or gantry but cant figure out where the exact problem is
Any suggestion please !!