Over 500 Total Lots Up For Auction at One Location - NJ 06/15

10391 RU:SYSTEMDOWN

Tags: PACS / DR / CR / Digital

(2)

CARLO BABA

10391 RU:SYSTEMDOWN

May 25, 2018 09:51

I have a FCR-GX-1 this is my first problem since I bought this scanner in 2008. It is still looks like brand new, and I don't want to ritier it just yet. Please, I check the service manual and I fund out this cod has to do with the arme that collecting dust from surface of the reading lamp.I really appreciate the help of some expert in this matter.

Log inor Register

to rate and post a comment

(50)
(7)

DIGITALXRAY

re: 10391 RU:SYSTEMDOWN

May 26, 2018 02:07

You may have a problem with the sensors. Check sensors or fuses.

Log inor Register

to rate and post a comment

(2)

CARLO BABA

re: re: 10391 RU:SYSTEMDOWN

May 30, 2018 09:40

Thank you for reply, would you know where are the fusees. I Have not seen any fuse in the machine or in the service manual. Again thank you