1800i 40ma error

Tags: CT, GE

Posted by an Anonymous User

1800i 40ma error

July 31, 2012 01:17

good day!!

i have a problem with this machine.(sytec1800i)single phase generator.

auto ma cal= success (all ma)including 40ma

during normal scan
60ma=ok
80ma=ok
100ma=ok
40ma=ER123 xg error
532 tube current accuracy error.
531 tube voltage accuracy error.

replace/test parts

hv tank= same error
heat rotor board=same error
cpu cmd syj=same error
new xray tube=same error
upgrade software on OC=same error

any advice is appreciated...

thank you in advance.

Log inor Register

to rate and post a comment

(10)
(1)

SUN YINGKAI

re: 1800i 40ma error

July 31, 2012 10:28

Are you adjust the Z-axis tube position?

Log inor Register

to rate and post a comment

(41)
(4)

john dismaya

re: re: 1800i 40ma error

August 01, 2012 11:33

hi sun,
this machine is working before.(without any adjustment of tube)
i also clean and try to swap the anode and cathode hv cable
but i got the same problem.

Log inor Register

to rate and post a comment

(10)
(1)

SUN YINGKAI

re: re: re: 1800i 40ma error

August 02, 2012 09:51

I think this may be the problem of TUBE. Are the two tubes same model? You can modify the EEPROM of XG. The manual said the very clearly. You look the value "D018",it is the parameter of 40mA, you can increase or decrease the value.

Log inor Register

to rate and post a comment

(10)
(1)

SUN YINGKAI

re: 1800i 40ma error

August 01, 2012 10:51

Please clean all HV cable plugs

Log inor Register