Over 500 Total Lots Up For Auction at One Location - NJ 06/15

3M 969 getting error 601 optics problem

(43)

(2)

Ernesto Guerra

3M 969 getting error 601 optics problem

October 31, 2002 10:02

I have a 3m 969 laser camera that gives me a 601 error. I have troubleshooted this problem and found that there is no signal caming back from one of the laser sensors in the optics assembly. Not been an expert I thing this is the problem. Any suggestions, schematics and parts are greatly appreciated. Thanks for your help. You can call me at 956-533-3863 or just reply to this email.

Log inor Register

to rate and post a comment

(8)

Guillermo Balletbo

Re: 3M 969 getting error 601 optics problem

November 01, 2002 06:11

Dear Ernesto,

The problem that you describe correspond to optics module failure and the solution for that, is to exchange it for a new one. We have an used
but operative module. If you are interested we can send you a quotation.
Sincerely.

Guillermo