Over 500 Total Lots Up For Auction at One Location - NJ 06/15

4024 Error, Siremobil Compact

(34)
(1)

JUAN PAULO REYES CASTILLO

4024 Error, Siremobil Compact

December 17, 2007 02:07

I have 4024 error in siremobil compact ,help me.i need password a service mode.
Best Regards
Paulo reyes [

Log inor Register

to rate and post a comment

(265)
(50)
(28)

Wayne Horsman

Re: 4024 Error, Siremobil Compact

December 19, 2007 06:28

Usually means bad Memoskop (video memory box)

Error E4024 Memoskop initialization timeout

Possible trouble: Host registered a timeout of the Memoskop during the
initialization of the system
Waiting time for the Memoskop reporting exceeded.

Possible help: Check Memoskop.
Check connection cable to the Memoskop.
Possibly Memoskop or D1 defective.

From - Ric Boule - Columbia Imaging Inc
410.499.9525
ric@columbiaimaging.com

Log inor Register

to rate and post a comment

(146)
(11)

Vic MayAl

re: 4024 Error, Siremobil Compact

December 08, 2014 06:11

Hello Juan Paulo,
I have work with a password that could be useful for you.
Mail me to vicmaal@yahoo.com.mx

Log inor Register

to rate and post a comment

Mohammed Abdullah

re: 4024 Error, Siremobil Compact

December 26, 2014 11:55

i have the manuals if you need it contact me

magmesme@gmail.com

regards

Log inor Register