4D problem in GE logiq 9

February 28, 2013 09:40

Hello,
the probe is 4D3C-L.
the probe works as a 2D probe only.
when I press the live button on the control panel,a message appears "4D probe reference.Check cable,reseat probe in socket"
the problem is in the machine?or the socket of the probe?or the motor of the probe?or what?
best regards

1 person liked this
Log inor Register

to rate and post a comment

(49)
(84)

zheng nichole

re: 4D problem in GE logiq 9

March 01, 2013 11:15

the problem in probe.
maybe in motor(if it's not move),in connector or other.
You should restart the machine and reseat probe in socket to check it first.
Nichole sales1@myprobe.cn

1 person liked this
Log inor Register

to rate and post a comment

(12)
(1)

NABIL Moustafa

re: 4D problem in GE logiq 9

March 02, 2013 09:10

Dear Eng,
First change te probe position if it the same problem check motor if it is working or not ,how to know change to other probe & back to probe again & put yor hand on the probe try to get the 4D mode
You can contact me at 00971553663662

Log inor Register

to rate and post a comment

(93)
(54)

mohamed elgenidy

re: re: 4D problem in GE logiq 9

March 02, 2013 03:01

thanks for ur reply.
I prefer the speaking to be here,so that,all the people can see reason of the problem.
I changed from this probe to another one,and returned to the 4D probe,and presses the "live" button,the problem is still available.
the motor isn't working using the 4D mode.

Log inor Register

to rate and post a comment

(22)
(1)

HUKAM CHAND

re: re: re: 4D problem in GE logiq 9

March 06, 2013 08:48

dear mohamed

the problem is not with your probe. the problem is with the high voltage supply responsible for running the 4D motor. Please replace that supply and it should work. This error is common in VOLUSON machines.

Please mail me if you have anything else. hukamsirhind@yahoo.com

Log inor Register

to rate and post a comment

(12)
(1)

NABIL Moustafa

re: 4D problem in GE logiq 9

March 02, 2013 09:10

Call me at 00971553663662

Log inor Register

to rate and post a comment

(12)
(1)

NABIL Moustafa

re: 4D problem in GE logiq 9

March 02, 2013 09:12

Contact me at info.ransy@gmail.com