For pre charge fail check your batteries. Your precharge capacitors C2 & C3 are not being charged by the batteries/battery charger. So the Technique Processor PCB does not see the +200VSNS signal reach approximately 4.5VDC during boot sequence (approximately 185 VDC across the precharge capacitors)
So check your batteries and battery charger PCB
REGULATOR FAIL:
filament voltage is out of range on mA regulator circuit. This is a non-fatal error.
The regulator fail error triggers the precharge error because the board fails to supply or regulate the proper 15V, 5V or 200V to the generator.
When the regulator fails, the Analog Interface or Generator Driver PCB cannot complete the precharge cycle.
1. Inspect voltage outputs of the xray regulator PCB
2. Check your batteries and battery charger board
3. If battery and charger PCB okay, check regulator PCB
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
MedicalTao
9600 regulator failures
May 26, 2026 10:10
I have an OEC with a regulator failure. Anyone with an ideal the cause? I am also getting a precharge error.
to rate and post a comment
CRAIG WAMBURA
re: 9600 regulator failures
May 28, 2026 10:05
For pre charge fail check your batteries. Your precharge capacitors C2 & C3 are not being charged by the batteries/battery charger. So the Technique Processor PCB does not see the +200VSNS signal reach approximately 4.5VDC during boot sequence (approximately 185 VDC across the precharge capacitors)
So check your batteries and battery charger PCB
REGULATOR FAIL:
filament voltage is out of range on mA regulator circuit. This is a non-fatal error.
The regulator fail error triggers the precharge error because the board fails to supply or regulate the proper 15V, 5V or 200V to the generator.
When the regulator fails, the Analog Interface or Generator Driver PCB cannot complete the precharge cycle.
1. Inspect voltage outputs of the xray regulator PCB
2. Check your batteries and battery charger board
3. If battery and charger PCB okay, check regulator PCB
Contact me for more details