Sorry for the delay in responding. These are voltages that are out of tolerance. You will need a new power supply.
The Accurus may use one-of-two power supplies. Power Supply II part number 202-1460-501 is assembled by Alcon, and its theory of operation is included here.
Power Supply III part number 202-1617-001 is a vendor-supplied product, and all information for this power supply is provided on the drawing included in section 6 of this manual.
Ophthalmic-focused expertise in YAG systems including repair, maintenance and calibration handled with precision by trained professionals. Request service • sales@laserlocators.com • 1-877-924-2020
Dimitris Afantitis
ACCURUS ERROR
April 01, 2019 09:12
907 and 903 error in accurrus vitrectomy system
please advice
to rate and post a comment
Dr.Phaco
re: ACCURUS ERROR
December 20, 2019 09:49
Sorry for the delay in responding. These are voltages that are out of tolerance. You will need a new power supply.
The Accurus may use one-of-two power supplies. Power Supply II part number 202-1460-501 is assembled by Alcon, and its theory of operation is included here.
Power Supply III part number 202-1617-001 is a vendor-supplied product, and all information for this power supply is provided on the drawing included in section 6 of this manual.
to rate and post a comment