October 21, 2005 10:44
I've an Acuson 128 XP10.
DTI is installed, but when I press the Alt-Color keys with the V2 Probe, the image is frozen, and I've to restart the system again.
Do you know where the problem may be?
It's an urgent issue, I've to solve during today, if possible.
I'd really appreciate your help
Thanks
walter
to rate and post a comment
Jim Carr
Re: Acuson 128 XP 10 - DTI - Problems
October 21, 2005 12:08
From our ultrasound system staff engineer (and former Acuson field engineer), Nicholas Mesagno:
It could be memex microcode or encryption codes. Maybe clearing out the BBM and re-installing codes.
to rate and post a comment
Walter Herman
Re: Acuson 128 XP 10 - DTI - Problems
October 21, 2005 02:14
Dear Jim/Nicholas.
We are just working on this.
If we don't have succeed with this, which is your suggestion ?
I'll let you know
Thanks for your cooperation
walter
to rate and post a comment
Walter Herman
Re: Acuson 128 XP 10 - DTI - Problems
October 21, 2005 03:26
Dear Jim/Nicholas
Nothing changed.We cleared out the BBM and re-installed codes.
To give you more info, DTI works in Spectral Mode, but it hangs in Color mode.
Thanks in advance
walter
to rate and post a comment
Emerson Freire
Re: Acuson 128 XP 10 - DTI - Problems
October 21, 2005 01:41
Walter,
What does it happens when you use normal color doppler, does the system hangs too?
Has the DTI ever worked on the system?
Emerson F. S. Freire
emerson_freire@hotmail.com
PS. If you need I will be online on MSN Messenger...
to rate and post a comment
Walter Herman
Re: Acuson 128 XP 10 - DTI - Problems
October 21, 2005 02:10
Dear Emerson
It works normal in color doppler
I don't know if DTI works before as I've just received the system , and I'm checking.
Walter
to rate and post a comment
Emerson Freire
Re: Acuson 128 XP 10 - DTI - Problems
October 21, 2005 02:53
Walter,
First of all check on the version report if you have a syscon 23 or higher and if you have:
v2.dti
v2
Then check in the configuration report to see if you have ART or Acoustic Res...
The DTI option needs all the following to work:
Boards: (All of them on the scanner)
- BBQ4
- IFD
- STG4 (At least)
- VDT3 (At least)
- The ART option needs to enabled. This will be displayed in the configuration report as Acoustic Res.
- You will need a syscon version 23 or higher...
Best Regards,
Emerson.
emerson_freire@hotmail.com
to rate and post a comment
Oral Unal
Re: Acuson 128 XP 10 - DTI - Problems
October 21, 2005 02:43
Walter
You will need to change the 3B50 (3.6V, 1.0Ah
Lithium) CVC batteries, reinstall the encryption strings
Good Luck
Necati Unal