Over 500 Total Lots Up For Auction at One Location - NJ 06/15

acuson 128 xp

(161)
(20)
(4)

Miguel Chavira

acuson 128 xp

November 05, 2005 01:13

hi ! i have a trouble whith ultrasound acuson 128 xp,
it isnt work fine in mode color flow, in mode pulsed dopler its work fine.
i run test PM and its find two fails

28F CDE2 YC CALIBRATION TEST failed
290 CDE2 COMPUSITE CALIBRATION TEST failed

some body help me with it?

thanks a lot!

Log inor Register

to rate and post a comment

Jim Carr

Re: acuson 128 xp

November 07, 2005 11:09

From our Systems Engineer, Rick Reed:

As I interpret him...color bad, PW ok. Probably FEC/FEB problem. 2 failures are related to VCR video, probably unrelated and a "don't care" fail.
rr

Log inor Register

to rate and post a comment

(161)
(20)
(4)

Miguel Chavira

Re: acuson 128 xp

November 07, 2005 04:36

ok thanks, the color flow have to fill artery or venus, red or blue , by only appear outside of flow..

thanks!!

Log inor Register

to rate and post a comment

Arturo Oropeza

Re: acuson 128 xp

November 10, 2005 03:54

hola Miguel yo estoy en la ciudad de aguascalientes y tengo un amigo especialista en acuson tal vez podria ayudarte