I have a Acuson 128/xp, boots up fine but there is a problem to get to service menu using test+code+shift+v, keyboard works fine and probably a test button works too,the 3.9V batteries are dead, just want to make sure this problem is not related to batteries.Please help!
Thanks.
Leon.
Comprehensive medical imaging support. We supply essential parts & maintenance for most OEMs, buy/sell quality used equipment, and offer mobile rentals globally. Contact us: info@lasoimaging.com
Leon Libgot
Acuson 128/xp menu problem
July 12, 2010 10:15
I have a Acuson 128/xp, boots up fine but there is a problem to get to service menu using test+code+shift+v, keyboard works fine and probably a test button works too,the 3.9V batteries are dead, just want to make sure this problem is not related to batteries.Please help!
Thanks.
Leon.
to rate and post a comment
Aram Arakelyan
Re: Acuson 128/xp menu problem
July 15, 2010 10:26
1. It is necessary will replace a battery
2 it is necessary to enter string codes
to rate and post a comment
Leon Libgot
Re: Acuson 128/xp menu problem
July 15, 2010 10:20
I tried to install a different CVC board with good batteries from another machine and still unable to get to service menu.
to rate and post a comment
Aram Arakelyan
Re: Acuson 128/xp menu problem
July 16, 2010 10:37
The string code it configuration systems depends on serial number
CVC it is not necessary directly will replace
to rate and post a comment
Leon Libgot
Re: Acuson 128/xp menu problem
July 16, 2010 11:05
I'll try to check cable to test button, but system still should go to the service menu in order to input strings.
to rate and post a comment
Aram Arakelyan
Re: Acuson 128/xp menu problem
July 17, 2010 10:08
service menu test+code+shift+v
to rate and post a comment
Aram Arakelyan
Re: Acuson 128/xp menu problem
July 17, 2010 06:46
string code is a matrix of hexadecimal numbers
13 x 8
CODE +SHIFT+TEST+V
Acuson 128 Diagnoctic code for units REV 26 +HIGHER
RESET+M.
• RESET+TEST (SAME TIME, LET go of RESET, THEN LET go of)
• S10BFBA (ENTER)
• 01 (ENTER)
• G262B600 (ENTER)
• PM (ENTER)
• KEYBORD =XF1
to rate and post a comment
Leon Libgot
Re: Acuson 128/xp menu problem
July 29, 2010 12:30
Problem was with the ribbon cable from test button pcb.