Acuson 128xp10 intermittent problem
May 18, 2004 04:24
When i turn on the system some time boot up, some time say initiallizing please wait then stop, some dosent recognize any probe some time does, and even the system boot up i never get the cw/pw or color function working.
please i need some help.
to rate and post a comment
Carlos Duran
Re: Acuson 128xp10 intermittent problem
June 18, 2004 05:49
hello:
probably you have poor contact problems in the lowest rack boards. clean connector of the pcbs in the rigth-back side )memory boards=.
in the upper rack, measure the voltage in the two battery in the third board, from back to front. if the voltaje is low, you may change this batteries.
good luck