Acuson -shutdown problems

March 29, 2016 04:37

Dear friends,

When I shutting down Siemens Antares it displays the window:

"System shutdown"
"Detecting network-Data transfer in progress"

Below it is blue bar that comes very slowly and then hangs (not responding). So I have to push and hold on/off button for 5 sec.to shutdown Antares.

I delete all patient data, but no result.

What do you recommend to do?

Log inor Register

to rate and post a comment

(165)
(67)
(2)

Sergey Zheleznyak

re: Acuson -shutdown problems

April 03, 2016 12:45

Hello,

Check all network job statuses and printer queue.
Also reinstallation of the software would definitely help.
Readout logs for the error descriptions.
email me s.v.zhel@gmail.com for further help

Regards

Log inor Register

to rate and post a comment

(10)
(1)

Luke Rzeznik

re: Acuson -shutdown problems

April 05, 2016 08:53

First. Make a SECTOR BY SECTOR copy of HDD. Then put it on another HDD (copy) and check if problem persist. Inform us.

Log inor Register

to rate and post a comment

(14)
(2)

MatLo

re: re: Acuson -shutdown problems

April 08, 2016 10:12

Thanks everyone to reply!
The problem is dissapeared by itself at the next day.
I suppose it was the problems with software.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: Medical Equipment Superstore Surplus Auctions July 14th

Medical Equipment Superstore closeout auctions take place on DOTmed.com. Bid today on 276 lots of medical equipment. Items can be inspected in person or via video call. Contact Rick at +1 (215) 669-3913 or info@medicalequipmentsuperstore.com

stats
advertisement

Upcoming Auction Events

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)