Over 150 Missouri Auctions End Today - Bid Now

Acuson Aspen Boot-Up Errors

(7)

(8)

Alex Loreti

Acuson Aspen Boot-Up Errors

January 13, 2004 04:07

We shipped a perfect working condition Acuson Aspen from our office to a hospital and it does not boot-up after delivery. The machine starts tp boot up but then gives the message: BOOT-UP ERRORS: UIC 0002 / SCP 002 PRESS ANY KEY TO CONTINUE. Any help will be appreciated. Thank you.

Log inor Register

to rate and post a comment

(3)

Duane Stork

Re: Acuson Aspen Boot-Up Errors

January 13, 2004 10:20

I would recomend that you turn the unit off and reseat all of the boards. Chances are something got jostled. Good luck
Duane Stork
Atlanta Ga

Log inor Register

to rate and post a comment

Jim Carr

Re: Acuson Aspen Boot-Up Errors

January 14, 2004 11:04

If Duane's suggestion doesn't do it, check the cables from the UIC to the ZAP. The ZAP is inside the Disk Drive Enclosure. The UIC is the User Interface Controller Board. Next I would check the cable from the UIC to the IOB (in the scan converter).

Log inor Register

to rate and post a comment

(1)

Piter Darichev

Re: Acuson Aspen Boot-Up Errors

January 27, 2005 02:00

Pleas help. Aspen gives the message:
"Software Sono-pac for this transducer has not been licensed"

Log inor Register

to rate and post a comment

(39)
(1)
(2)

Emerson Freire

Re: Acuson Aspen Boot-Up Errors

January 29, 2005 06:22

If the problems persist, check under the keyboard / User Interface Board for any shorts....


And answering Alex question: If the transducer was working before, due to a problem on a battery you lost the encryption strings - if this is the case, contact Siemens and ask them a copy of it...
If it is a new transducer it may not be enabled...

Log inor Register

to rate and post a comment