Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

Acuson Sequoia 512 problem

(20)
(6)

Alexander Ershov

Acuson Sequoia 512 problem

October 03, 2011 11:00

Hi all!

Anybody have service manual Acuson Sequoia 512? I need service password for pass service level. Device is sometimes power off, and LED CODC OVP is light. My e-mail ershov777@nm.ru.

Best regards, Alexander Ershov

Log inor Register

to rate and post a comment

(35)
(2)
(3)

Robert Celmer

Re: Acuson Sequoia 512 problem

October 03, 2011 06:10

This is the over voltage protection light. Check incoming power to make sure it is the correct voltage, if your area has power spikes you should put a UPS system on the power and plug the system into it. if the power supply from your outlet is stable and does not spike you need a new power supply for your system

Log inor Register

to rate and post a comment

(23)

Kir kavra

re: Re: Acuson Sequoia 512 problem

September 15, 2015 06:20

MY MACHINE STARTS AND AFTER BOOTING UP IT SHUTDOWN AND RESTARTS AGAIN AFTER A FEW MINUTES... CAN YOU PLEASE TELL ME WHY? MY EMAIL IS : MUHC08@GMAIL.COM

dONALD

Log inor Register

to rate and post a comment