Please help.
At once I want to apologize for my English. I want to ask for you the help concerning X-150.
Malfunction TI has been revealed - this block incorrectly defines a submitted high voltage from the power unit. If a tester of pressure I see 100 Volt and +100 Volt stably TI defines them as -112 Volt and +114 Volt and naturally at once switches off the X-150 without possibility to come into the service menu.
For couple of days I have studied TI have drawn the electric scheme of module TI.
I replace on the TI module:
IRZ9210, IR210, LM339 (2) and I cant find to replace DG604W.
But it dont work:(
For now I need to know how to calibrate the measured high voltage by module TI. The calibration files is available in system are channels 32 and 35 (look more low).
How I can launch ZDG calibration? I cant enter to the shell:(
Whether probably to make so that tests on these two channels were not spent?
I.e. a tester of pressure I see-100 Volt and +100 Volt, and TI measures it with huge errors: (
One time I always replace this TI module but dont want to do this another one:(
I want to try repair it or calibrate.
I give information of repair TI module if it happens with your help.
Thx.
START DIAG A2D TEST 07/07/2010 13:30:17
Chan=35 Volts=112.36 ChanName="100VF - Pos 100 Volt Scan Head PWR" BRD=
Chan=32 Volts=-114.47 ChanName="-100VF - Neg 100 Volt Scan Head PWR" BRD=
And part of calibrating file:
START OFFSET CAL 03/19/2007 15:37:53
Median=2.0 Vmid=0x4F
Thx for your reply! But I have service manual for x150 and there are no instruction how to calibrate it.
If you have that instuction pls give it to me.
DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com
Kasparoff Vital
ACUSON X150 HELP!!!!!!!!!!!!!
October 06, 2010 12:40
Hello from Russia!
Please help.
At once I want to apologize for my English. I want to ask for you the help concerning X-150.
Malfunction TI has been revealed - this block incorrectly defines a submitted high voltage from the power unit. If a tester of pressure I see 100 Volt and +100 Volt stably TI defines them as -112 Volt and +114 Volt and naturally at once switches off the X-150 without possibility to come into the service menu.
For couple of days I have studied TI have drawn the electric scheme of module TI.
I replace on the TI module:
IRZ9210, IR210, LM339 (2) and I cant find to replace DG604W.
But it dont work:(
For now I need to know how to calibrate the measured high voltage by module TI. The calibration files is available in system are channels 32 and 35 (look more low).
How I can launch ZDG calibration? I cant enter to the shell:(
Whether probably to make so that tests on these two channels were not spent?
I.e. a tester of pressure I see-100 Volt and +100 Volt, and TI measures it with huge errors: (
One time I always replace this TI module but dont want to do this another one:(
I want to try repair it or calibrate.
I give information of repair TI module if it happens with your help.
Thx.
START DIAG A2D TEST 07/07/2010 13:30:17
Chan=35 Volts=112.36 ChanName="100VF - Pos 100 Volt Scan Head PWR" BRD=
Chan=32 Volts=-114.47 ChanName="-100VF - Neg 100 Volt Scan Head PWR" BRD=
And part of calibrating file:
START OFFSET CAL 03/19/2007 15:37:53
Median=2.0 Vmid=0x4F
Chan=32 Offset=3
Chan=35 Offset=2
END OFFSET CAL
to rate and post a comment
yuan bo
Re: ACUSON X150 HELP!!!!!!!!!!!!!
October 07, 2010 12:24
i think you can calibrate the machine according the instruction of the service manual.
to rate and post a comment
Kasparoff Vital
Re: ACUSON X150 HELP!!!!!!!!!!!!!
October 07, 2010 09:00
Hi!
Thx for your reply! But I have service manual for x150 and there are no instruction how to calibrate it.
If you have that instuction pls give it to me.
to rate and post a comment
Kasparoff Vital
Re: ACUSON X150 HELP!!!!!!!!!!!!!
February 20, 2012 10:09
I found solution for this trouble. If you have it send me e-mail.
to rate and post a comment
bobochana
re: Re: ACUSON X150 HELP!!!!!!!!!!!!!
April 26, 2018 03:24
You do? Can you you tell me what is?
to rate and post a comment
BillWard
re: Re: ACUSON X150 HELP!!!!!!!!!!!!!
August 19, 2015 04:09
would you please send me a copy of the manual? Thank you. / Bill
to rate and post a comment
BillWard
re: Re: ACUSON X150 HELP!!!!!!!!!!!!!
August 22, 2015 09:39
are you willing to share? I need a manual. Thank you. / Bill bward@freedomimaging.com.
to rate and post a comment
bobochana
re: Re: ACUSON X150 HELP!!!!!!!!!!!!!
April 26, 2018 03:24
Hi
Do you have a Acuson X150? If you do please send to my mail ilidio@salutec.pt
to rate and post a comment