Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Acuson X300 Error

Tags: Ultrasound, SIEMENS

(4)
(1)

Dennis Him

Acuson X300 Error

March 31, 2026 11:16

Hi, we have a ACUSON X300 with the following Error:

The System is shutting down automatically to prevent permanent damage, please contact Siemens Service.

I cant see a Error code - who can help? Whats the Issue?

Equipment: Siemens ACUSON X300 Premium Edition

I found 2 faulty Caps somebody any advice before i start changing parts?

Images:
https://ibb.co/album/v1GnCs

1 person liked this
Log inor Register

to rate and post a comment

(1)

Alex Shan

re:Acuson X300 Error

March 31, 2026 06:05

how about selftest during startup?. it is pass? or error popup during bootup? try reload software

Log inor Register

to rate and post a comment

(4)
(1)

Dennis Him

re: re:Acuson X300 Error

April 01, 2026 10:20

how to reload software? I found out that the DCDC Converter does not have 100v but the PSU have 100v so something on the DCDC Converter

Log inor Register

to rate and post a comment

(4)
(1)

Dennis Him

re: re:Acuson X300 Error

April 01, 2026 10:20

how to reload software? it passes selftest, the error comes when the software already loaded, and i cannot access service menu because that pos shuts down

Log inor Register

to rate and post a comment

(1)

AlexShan

re: re: re:Acuson X300 Error

April 01, 2026 02:23

Sounds like your PSU might have an issue. Hopefully it can be replaced soon.

Log inor Register

to rate and post a comment

(8)

Hermann Gilly

re: Acuson X300 Error

April 01, 2026 10:20

I am sorry that I cannot really help you but I have experienced the same error with Siemens Acuson 150. No helpful answer of the company. I replaced the power supply - no change. Sometimes (after long intervals with the system turned off) the device started correctly but failed again after about 30 mins. Opening the system´s enclosure to avoid heating up, it up also did not succeed. Unfortunately Siemens does not provide a service manual. Eventually you could check the diagnostic led indicators the test pins of the power supply. I may send the details via seperate mail.

Log inor Register

to rate and post a comment

(4)
(1)

Dennis Him

re: re: Acuson X300 Error

April 01, 2026 02:23

your device is definitely getting hot somewhere - only way is to maybe use a thermal camera to check

Log inor Register

to rate and post a comment

stats
DOTmed text ad

BK and other Urological Systems available. BK Probe repair available!!!

Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net

DOTmed Certified stats
advertisement

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)