Hello ,
I have an issue with a Siemens Acuson X300 PE.
The unit boot up normally and arrives in the state that you are ready to start the examination.
In the upper left of the monitor it recognises normally the connected probes but the problem is that you cannot see the cone of any conected probe.
even if you maximice the gain to 100%. But when you freeze the image the cone arrears cuted is sections of black and grey.
I have stored image to show you.
Has anyone had same more or less problem.
Can you help me please ?...
Hello Jose and thank you.
I assume it also but the difficult part of this is that the defect it is not constant. For example now is working normally.
Even so, I must choose between RC and TX board.Correct ?
The internal test that the unit has in service window says is all ok.
Could this be of an external problem ? Bad quality on the wall outlet or due to a malfuntion UPS ? (it is connected to a ups system)
If as you mentioned the problem is intermittent, then you will have to wait until it becomes permanent. One thing you can do in the meantime and that is take out the tx and rc and clean their contacts, then reset them again, it might be a case of false contact.
I have many problems with X300 PE, but I have not service manual of it.
If you have, please send me a copy of it!
My e-mail: biomedicalengineervn@gmail.com
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
Zeratul
Acuson X300 Image problem
July 22, 2013 10:19
Hello ,
I have an issue with a Siemens Acuson X300 PE.
The unit boot up normally and arrives in the state that you are ready to start the examination.
In the upper left of the monitor it recognises normally the connected probes but the problem is that you cannot see the cone of any conected probe.
even if you maximice the gain to 100%. But when you freeze the image the cone arrears cuted is sections of black and grey.
I have stored image to show you.
Has anyone had same more or less problem.
Can you help me please ?...
Thank you
to rate and post a comment
Jose Pacheco
re: Acuson X300 Image problem
July 22, 2013 04:37
that problem is at the front end, it can be the front end controller or a channel board(or transmitter / receiver.
to rate and post a comment
Zeratul
re: re: Acuson X300 Image problem
July 23, 2013 08:39
Hello Jose and thank you.
I assume it also but the difficult part of this is that the defect it is not constant. For example now is working normally.
Even so, I must choose between RC and TX board.Correct ?
The internal test that the unit has in service window says is all ok.
Could this be of an external problem ? Bad quality on the wall outlet or due to a malfuntion UPS ? (it is connected to a ups system)
to rate and post a comment
Jose Pacheco
re: re: re: Acuson X300 Image problem
July 23, 2013 01:55
If as you mentioned the problem is intermittent, then you will have to wait until it becomes permanent. One thing you can do in the meantime and that is take out the tx and rc and clean their contacts, then reset them again, it might be a case of false contact.
to rate and post a comment
Zeratul
re: re: re: re: Acuson X300 Image problem
July 23, 2013 04:30
How do I reset them ?
Will this affect the unit ?
to rate and post a comment
Biomedical Engineer
re: re: Acuson X300 Image problem
April 17, 2014 01:16
Hello Jose,
I have many problems with X300 PE, but I have not service manual of it.
If you have, please send me a copy of it!