ADAC Dual Head Pegasys problem

May 11, 2007 08:05

We have ADAC Dual Head Gamma Camera with Pegasys Software Version 3.4

We are trying to load Cedars AutoSPECT option using Toshiba CD ROM Drive.

I have tried the drive ID at 0, 1,2 and 5. It does not work. The Tape and HD are ID at 3 and 4.

Can some one help
Thank you in advance

Harkesh Singh
917 856 5211
info@ahtiny.com

Log inor Register

to rate and post a comment

(75)
(8)
(7)

JIFFY PAUL AKKARAKARAN

Re: ADAC Dual Head Pegasys problem

May 11, 2007 04:12

Hi,
Have you checked the SCSI termination setting? Check to which device the termination is made active.

Jiffy Paul


Log inor Register

to rate and post a comment

Lee Kelly

Re: ADAC Dual Head Pegasys problem

May 13, 2007 07:53

Harkesh, the ADAC software is looking for a specific scsi id when loading software or options. The id would be the same as the cdrom drive. To determine the scsi id you shutown and at the ok prompt you type show-scsi and it will show you the cdrom's id. you then remove the cdrom and put your other device in its place. I believe the id is set to 6. You can call American Allied Nuclear, Inc. and our support is free.
Lee Kelly
866-782-2687

Log inor Register

to rate and post a comment

(6)

(6)

Harkesh Singh

Re: ADAC Dual Head Pegasys problem

May 14, 2007 11:40

Lee, Thank you. It worked at ID 6. I was able to load the software.
I will call you today to spaek with you.
Thank you

Harkesh Singh 917 856 5211