January 24, 2019 09:01
Hello. GE Innova 2000 giving the following error: 42C5 Maximum tube hausing heat units reached.
The Chiller is working and in good condition. Temp is 20 degrees Celcius. The voltages across the wires also hasnât problem. Have anyone faced such problem? Help please.
muhammad.koziev@gmail.com
to rate and post a comment
Michael Fisher
re: Advantx Generator
January 26, 2019 09:40
My first assumption is if all the cooling system is operating correctly then the tube heat sensor is most likely defective. You should be able to jumper across the terminals and verify. System will work but eventually you will need a new tube.
to rate and post a comment
Irshod Yakubi
re: re: Advantx Generator
January 27, 2019 12:32
I fixed this problem but got another problem. The AdvantX is giving 42C4 Tube anode not rotating. Replacing TIRC didnât solve the problem. All cables are in good condition. Can the X-ray tube be broken?
Sincerely.