Hello. GE Innova 2000 giving the following error: 42C5 Maximum tube hausing heat units reached.
The Chiller is working and in good condition. Temp is 20 degrees Celcius. The voltages across the wires also hasn’t problem. Have anyone faced such problem? Help please.
muhammad.koziev@gmail.com

Log inor Register

to rate and post a comment

(64)
(12)
(3)

Michael Fisher

re: Advantx Generator

January 26, 2019 09:40

My first assumption is if all the cooling system is operating correctly then the tube heat sensor is most likely defective. You should be able to jumper across the terminals and verify. System will work but eventually you will need a new tube.

1 person liked this
Log inor Register

to rate and post a comment

Irshod Yakubi

re: re: Advantx Generator

January 27, 2019 12:32

I fixed this problem but got another problem. The AdvantX is giving 42C4 Tube anode not rotating. Replacing TIRC didn’t solve the problem. All cables are in good condition. Can the X-ray tube be broken?

Sincerely.
whatsapp: +992927641117

Log inor Register

to rate and post a comment

Irshod Yakubi

re: re: Advantx Generator

January 27, 2019 12:32

I fixed this problem but got another problem. The AdvantX is giving 42C4 Tube anode not rotating. Replacing TIRC didn’t solve the problem. All cables are in good condition. Can the X-ray tube be broken?

Sincerely.
whatsapp: +992927641117

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Probo Medical | We Buy, Repair, and Sell New and Used Ultrasound Equipment

We provide sales, service, rental, and repair for ultrasound systems, probes, and C-arms. We have the largest inventory in the world. Contact us about our special offers on certified refurbished equipment! Call: 317-759-9210 Email: sales@probomedical.com.

stats

Upcoming Auction Events

Cbrx2sales
TIMED
Jun 05, 2026 (157 lots)