Dears i have 6111 error Code, in advantx lca, Omega table, CPU in positioner cabinet restart and fail led red is up
someone has experience of this problem? and is there a procedure to determine the failure?
Regards
PAULO
One of the following happened: 1) POS was not configured on a mode configure because
the POS was never marked online. 2) No interlock reply was received from the Positioner
when attempting fluoro exposures. 3) No reply to the Configure for Acquisition command
was received from the Positioner when attempting fluoro exposures. 4) An error status was
received in reply to Configure for Acquisition command from the Positioner when
attempting fluoro exposures.
RECOMMENDED ACTIONS
1. Ensure that the Positioner is online and functional.
2. Cycle console power so that it can be marked online in the ASC.
From preventive maintenance through comprehensive cryogen services, BMI-PMI is your trusted partner for new/reconditioned sales, installation, and service of both facility-based and mobile MRI solutions. www.brownsmedicalimaging.com / 800-701-9729
JUAN PAULO REYES CASTILLO
Advantx lca 6111 error code
November 20, 2018 09:54
Dears i have 6111 error Code, in advantx lca, Omega table, CPU in positioner cabinet restart and fail led red is up
someone has experience of this problem? and is there a procedure to determine the failure?
Regards
PAULO
to rate and post a comment
Francesco Didonato
re: Advantx lca 6111 error code
November 21, 2018 10:01
Hi you have problem on CPU board RAM lost parameter voi must change RAM and put in parameter by handle
Regards
to rate and post a comment
Johnny Barreto
re: Advantx lca 6111 error code
November 21, 2018 10:21
RECOMMENDED ACTIONS
1. Ensure that the Positioner is online and functional.
2. Cycle console power so that it can be marked online in the ASC.
to rate and post a comment
Maurice Dilick
re: Advantx lca 6111 error code
November 21, 2018 10:21
One of the following happened: 1) POS was not configured on a mode configure because
the POS was never marked online. 2) No interlock reply was received from the Positioner
when attempting fluoro exposures. 3) No reply to the Configure for Acquisition command
was received from the Positioner when attempting fluoro exposures. 4) An error status was
received in reply to Configure for Acquisition command from the Positioner when
attempting fluoro exposures.
RECOMMENDED ACTIONS
1. Ensure that the Positioner is online and functional.
2. Cycle console power so that it can be marked online in the ASC.