Johnny Barreto

ADVANTX-E comunication problem with DLX3

February 12, 2015 04:01

I'm having trouble to comunicate the Advantx system with the DLX-3 module, Not aquisition ready indicator on the DLX flat panel. The service manual direct me to run a test for GEIM board of the DLX, using a GELOOP CABLE, that I don't have, to connect it between two BNC conectors. Does anyone know if that GELOOP cable is just a short coaxial cable? Or it's a special one?

Log inor Register

to rate and post a comment

(10)
(3)

Natthakit Prongchai

re: ADVANTX-E comunication problem with DLX3

February 13, 2015 10:17

Which system are you working on? Innova 2000 or else

Log inor Register

to rate and post a comment

(128)
(10)
(15)

Johnny Barreto

re: re: ADVANTX-E comunication problem with DLX3

February 13, 2015 03:36

The system is an Advantx LCA+.

Log inor Register

to rate and post a comment

(128)
(10)
(15)

Johnny Barreto

re: re: ADVANTX-E comunication problem with DLX3

February 13, 2015 03:37

It's an Advantx LCA+. Titan sw verson G7.11

Log inor Register

to rate and post a comment

Charlie Britt

re: re: re: ADVANTX-E comunication problem with DLX3

February 14, 2015 01:25

hi, I think you have a problem with the nvram of mvme167.Is this the original problem? I mean, have you reload dlx software, or have you replaced the vmme167?
You have to check the boot of nvme167

Log inor Register

to rate and post a comment

Joseph A. Nelsen

re: ADVANTX-E comunication problem with DLX3

February 14, 2015 01:20

Several things to check. One is there may be a double termination on the coaxial network cable. Put a bnc T connector at one port and measure, with system off, the ohmic value seen at the open side of the T connector. It should be 50 ohms. less means douible terminated somewhere. you should be able to chase it down by sequencely removing eash bnc coas cable one at a time.

second, the Adv E kernal module has an auto termination built into the front panel bnc connector. It is a POS. It can cause an unterminated condition at the Adv E box. test that by putting a terminator on the outside of the box and see how it behaves, or with system off, measure the ohmic value there. 50 ohms ok, other not so much. leaving the outboard terminator on the front precludes closing the large cabnet door. I used to open the kernal box and remove the mini coas from the front bnc with the self termination and add ther terminator resistor internally on the end of the coas, under/behind the kernal pannel. At least that what I feel your problem sounds like.At least if my memory serves after about 20 years! Good luck!

Log inor Register

to rate and post a comment

(120)
(15)
(15)

Henrique Gomes

re: re: ADVANTX-E comunication problem with DLX3

March 20, 2015 08:56

what´s your email, Joseph?

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: ADVANTX-E comunication problem with DLX3

February 16, 2015 05:22

no just a short coaxial cable to complete the circuit

Log inor Register

to rate and post a comment

(128)
(10)
(15)

Johnny Barreto

re: re: ADVANTX-E comunication problem with DLX3

February 23, 2015 08:20

Dear Lenny, what kind of cable is it? Could I make anything to replace?

Log inor Register

to rate and post a comment

(128)
(10)
(15)

Johnny Barreto

re: ADVANTX-E comunication problem with DLX3

February 23, 2015 08:20

Still cannot comunicate!! I checked the VME167, all passed. Measured the 50 Omh at BNC connectors, in fact, I got 18 Ohm, I removed avery connector at a time, and all others than the ASC at once, but only at this time I got the 50 Ohms.
"The DX node is offline" message is given in diagnostics and service mode when trying to connect with the DLX unit.

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: ADVANTX-E comunication problem with DLX3

February 23, 2015 01:48

when you checked the VME you che checked all the environment settings?

Log inor Register

to rate and post a comment

(128)
(10)
(15)

Johnny Barreto

re: ADVANTX-E comunication problem with DLX3

February 24, 2015 10:28

Thank you all guys! Problem solved. The fiber optic cables that carry the GE Bus were badly labeled and misconnected. I just swapped them, and bingo! The blue to grey, the grey to blue and worked fine!

Log inor Register

to rate and post a comment