AGFA 5302_RF2 DEFECTIVE

October 12, 2012 01:28

AGFA 5302 DISPLAYS ERROR RF2 AT STARTUP.
CAN ANYONE ASSIST IN SOLVING THIS PROBLEM

Log inor Register

to rate and post a comment

(71)
(8)

TOM CHARLES

re: AGFA 5302_RF2 DEFECTIVE

October 12, 2012 04:43

check RF TAG READER that is defective in upper tray or you have do something wrong like you place an OLD film package check these two things if not feel free to contact me at tomcharles41@yahoo.com

1 person liked this
Log inor Register

to rate and post a comment

(33)
(3)

polisetty suresh babu

re: re: AGFA 5302_RF2 DEFECTIVE

April 04, 2014 02:14

hello sir, my agfa 5302 showing "system failure." any help

Log inor Register

to rate and post a comment

(9)

gajhendran gajhen

re: re: re: AGFA 5302_RF2 DEFECTIVE

April 07, 2014 08:44

please kindly check your hard disk.thank you

Log inor Register

to rate and post a comment

(33)
(3)

polisetty suresh babu

re: re: re: re: AGFA 5302_RF2 DEFECTIVE

April 07, 2014 04:19

how to check ds5302 hard disk sir?

Log inor Register

to rate and post a comment

(83)
(42)
(1)

Mohammad Ali Haidary

re: re: AGFA 5302_RF2 DEFECTIVE

April 07, 2014 08:49

Dear Polisetty Suresh Babu,
would you plese send me the Agfa 5302 Service manual.
i will also contact your email that you have provided and will ask for some supports.
BR Ali

Log inor Register

to rate and post a comment

(83)
(42)
(1)

Mohammad Ali Haidary

re: re: AGFA 5302_RF2 DEFECTIVE

April 08, 2014 07:34

Dear Tom Charles,
would you plese send me the Agfa 5302 Service manual.
i will also contact your email that you have provided and will ask for some supports.
BR Ali

Log inor Register

to rate and post a comment

(23)
(7)

Alexander Ethan

re: re: re: AGFA 5302_RF2 DEFECTIVE

April 09, 2014 03:31

I have a service manual, if you need it email me,
engineer_7771@yahoo.com

Log inor Register

to rate and post a comment

(83)
(42)
(1)

Mohammad Ali Haidary

re: re: re: re: AGFA 5302_RF2 DEFECTIVE

April 10, 2014 08:58

Dear Alexander Ethan,
Thanks for your response i have email you.
please send me acopy of pdf file.
BR
Ali

Log inor Register

to rate and post a comment

(22)
(4)

Fedor Mikolakhin

re: AGFA 5302_RF2 DEFECTIVE

October 12, 2012 06:20

Agree with Tom Charles.
First, try different film pack or current pack in different tray. If nothing changes, it must be something with rf-tag reader part.
For troubleshooting you should check cables from RF-tag reader (underneath tray) for loosen connections or even swap upper and lower trays: in DS5302s that i've seen they was identical.
Never faced firmware failure, it's barely possible, i think.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats
advertisement

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)