AGFA 5500 Error

September 08, 2017 10:46

Please, we have two units of AGFA 5500. On boot up, first one says, “Close the print section.” To the best o my knowledge, the print section is not open. What could be the problem?

The second machine gives an error message “System failure” on boot up. Could anyone help with possible solutions? Email: rainbowtroutng@gmail.com

Log inor Register

to rate and post a comment

(25)
(5)
(2)

Marwan Kumsie

re: AGFA 5500 Error

September 11, 2017 09:40

Erfan,
What I remember that there is a mechanism with belt at the lower door or at the print section and you must look there.This belt is not in position to activate some switch so it gives you an error .
Regards

Log inor Register

to rate and post a comment

(16)

Efman

re: re: AGFA 5500 Error

September 15, 2017 03:11

Thanks. I really appreciate

Log inor Register

to rate and post a comment

(410)
(57)

CRAIG WAMBURA

re: AGFA 5500 Error

September 11, 2017 02:58

should be a sensor problem.

For second one, perform an emergency install

Log inor Register

to rate and post a comment

(16)

Efman

re: re: AGFA 5500 Error

September 15, 2017 03:12

Would check. Thank you so much

Log inor Register

to rate and post a comment

stats
DOTmed text ad

GE, Siemens, Philips, & Toshiba/Canon imaging parts, training & support!

DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com

DOTmed Certified stats
advertisement

Upcoming Auction Events

Cbrx2sales
TIMED
Jun 05, 2026 (157 lots)