Ampronix, a Top Master Distributor for Sony Medical, provides Sales, Service & Exchanges for Sony Surgical Displays, Printers, & More. Rely on Us for Expert Support Tailored to Your Needs. Email info@ampronix.com or Call 949-273-8000 for Premier Pricing.
javed khan
agfa adc compact error 20507
January 05, 2015 08:53
Hello all.
I have adc compact cr. an error 20507, (digitaizer not ready please wait)Please someone help me. Thanks
to rate and post a comment
Rebel Cook
re: agfa adc compact error 20507
January 05, 2015 03:57
Laser power is below or equal to shutdown. Reason: Machine was switched on and laser has not reached its operating temp, bad laser, laser beam not visible by pin diode, galvanometer has no power or sweep signal.
to rate and post a comment
HyderAli12
re: agfa adc compact error 20507
January 05, 2015 06:24
20507
E_SCN_LASER_WAIT_SHTDN
Meaning : Laser power is below or equal the shutdown level
during initial starting phase. Machine status is "locked".
Reason : - Machine was just switched on and the laser has not yet
its operating temperature.
- Bad laser.
- Laser beam not visible by PIN diodes.
- Polygon gets no power or sweep signal.
Cure : This is not yet an error condition.
If the laser power does not become ok or warning during
20 minutes, an error is displayed.
to rate and post a comment
Vo Quoc
re: agfa adc compact error 20507
January 08, 2015 08:51
Hi Jave
The root issue maybe causing from the way of laser beem runing,
You can open out the scan unit , power on ADC compact and wait for ADC start up, while selftest processing you can see laser beam appear from the laser tube to galvo, to optic, .....to PMT
some case, clean some mirror also can fix the issue
Good Luck to you
Bao
to rate and post a comment