Djamel Boutemine
AGFA ADC COMPACT Error 22497
June 15, 2016 07:47
Hello,
We are having trouble with the CR system AGFA ADC compact.
The Error code is 22497.
could someone kindly send us the description of this error?
Kindest Regards
Djamel.
to rate and post a comment
Haider Ali
re: AGFA ADC COMPACT Error 22497
June 15, 2016 06:01
22497
E_CAM_INIT_TOUNCLAMP
Reason : Cassette module : Timeout of stepper drive clamping.
Action: Home positioning.
Meaning : - Drive doesn't reach home pos. because of any mechanical problems.
- No stepper power supply.
- Problems with stepper motor.
- Problems with home pos. light barrier.
- Problems with stepper node.
Cure : - Check mechanics.
- Check stepper power supply.
- Check motor (connected ?).
- check/replace home pos. light barrier and cable.
- Check/replace stepper node.
to rate and post a comment
Djamel Boutemine
re: re: AGFA ADC COMPACT Error 22497
June 17, 2016 09:54
Thanks for your help
Much appreciated
to rate and post a comment
Haider Ali
re: re: re: AGFA ADC COMPACT Error 22497
June 17, 2016 11:44
you are wellcome