It is an error on the rotation of the cassette handler. It is either not reaching the home position or the light barrier needs adjusted or the light barrier sensor is bad. It is located on the back wall of the CR reader. It could also be the flag on the rotational robot itself, it could be broken off or misaligned.
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
jesus paramo
AGFA COMPAC PLUS CR.
March 24, 2012 01:43
I HAVE AGFA COMPACT PLUS CR AND POWER ON THE DIGITIZE IT COMES ERROR CODE 22871, ANY BODY KNOWS THIS PROBLEM . ANYBODY HAVE THE MANUALS TECNICS.
THANK YOU. royparamo@yahoo.com.mx
to rate and post a comment
Juan de la Cruz Martinez
Re: AGFA COMPAC PLUS CR.
March 24, 2012 07:46
Enter the service menu and select:
4 Show error
1 Explain error code
to rate and post a comment
jesus paramo
Re: AGFA COMPAC PLUS CR.
March 25, 2012 12:25
DO YOU HAVE PDF THE TECNIC MANUAL , IF YOU SEND OR BOUGHT ME ?
THANK YOU.
jESUS R OGELIO PARAMO .
to rate and post a comment
Juan de la Cruz Martinez
Re: AGFA COMPAC PLUS CR.
March 25, 2012 11:07
I think I do, let me find it in the office tomorrow
to rate and post a comment
Arbjan Rusi
Re: AGFA COMPAC PLUS CR.
March 28, 2012 07:32
can you senda copy of this documentation to me?
thanks
tiralb123@gmail.com
to rate and post a comment
J Adams
Re: AGFA COMPAC PLUS CR.
March 25, 2012 12:31
It is an error on the rotation of the cassette handler. It is either not reaching the home position or the light barrier needs adjusted or the light barrier sensor is bad. It is located on the back wall of the CR reader. It could also be the flag on the rotational robot itself, it could be broken off or misaligned.
to rate and post a comment
shahed murad
re: Re: AGFA COMPAC PLUS CR.
January 27, 2016 10:37
hello brain thoms
i need the agfa compact plus cr service manual. please help
my email is shahed_murad@hotmail.com
to rate and post a comment