I have an digitizer Agfa CR 15-X type 5151 300, and It been down for several days, when I start the system I get the following code: ARC2805, I need your help please.
re: Agfa CR 15-X Defective with error code ARC2805
March 02, 2021 10:10
From your error it looks like the PMI board has failed or the power board.
Check the cable from the Power Board to the PMI, if necessary replace the cable.
Check the Power Board, it may have failed.
Check the fuses and output voltages on the power board.
If fuses are okay but output voltages are not then replace the power board. You cannot exchange the power board with another unit, they are branded to your specific unit at the time of turn on.
Or your PMI Board has failed, replace the PMI Board.
You cannot exchange the PMI Board with another system, they are branded once turned on to your system.
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
Abdelkader Guesmia
Agfa CR 15-X Defective with error code ARC2805
March 01, 2021 10:50
I have an digitizer Agfa CR 15-X type 5151 300, and It been down for several days, when I start the system I get the following code: ARC2805, I need your help please.
to rate and post a comment
Ronald Swapp
re: Agfa CR 15-X Defective with error code ARC2805
March 02, 2021 10:09
Why don't you http into the digitizer and it will tell you what the error is.
to rate and post a comment
Ronald Swapp
re: Agfa CR 15-X Defective with error code ARC2805
March 02, 2021 10:10
From your error it looks like the PMI board has failed or the power board.
Check the cable from the Power Board to the PMI, if necessary replace the cable.
Check the Power Board, it may have failed.
Check the fuses and output voltages on the power board.
If fuses are okay but output voltages are not then replace the power board. You cannot exchange the power board with another unit, they are branded to your specific unit at the time of turn on.
Or your PMI Board has failed, replace the PMI Board.
You cannot exchange the PMI Board with another system, they are branded once turned on to your system.
to rate and post a comment