Agfa CR 25.0 error

Tags: PACS / DR / CR / Digital

(4)
(1)

Salvatore Tovarich

Agfa CR 25.0 error

November 23, 2017 05:28

Could you help me with the error 22497 Agfa CR 25.0 and send service manual please thank you very much for your help

1 person liked this
Log inor Register

to rate and post a comment

(432)
(83)
(3)

juan pablo amado

re: Agfa CR 25.0 error

November 25, 2017 10:10

22497
E_CAM_INIT_TOUNCLAMP
Reason : Cassette module : Timeout of stepper drive clamping.
Action: Home positioning.
Meaning : - Drive doesn't reach home pos. because of any mechanical problems.
- No stepper power supply.
- Problems with stepper motor.
- Problems with home pos. light barrier.
- Problems with stepper node.
Cure : - Check mechanics.
- Check stepper power supply.
- Check motor (connected ?).
- check/replace home pos. light barrier and cable.
- Check/replace stepper node.

1 person liked this
Log inor Register

to rate and post a comment

(16)
(1)

medicalimagecare

re: Agfa CR 25.0 error

November 25, 2017 10:17

Dear sir May be Light Barior senser or moter stepper board are defective are loose conection
my whatsapp no is +918709759220

1 person liked this
Log inor Register

to rate and post a comment