AGFA CR 30-X error MESS263

September 05, 2016 10:34

Hello all.
I need help with a random error showed in the NX software with a CR 30-X digitizer, error code MESS263. I have a manual for the NX software but i can't find this error on it.

Thanks in advanced.

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: AGFA CR 30-X error MESS263

September 06, 2016 04:05

how full is the database? I have a note where I had the same error 263 and it turned out to be a database issue. I ended up deleting the database to correct