Dear all
I have a problem at agfa cr 35 image .
A horizantal blure line appear at any ip size.
This line appeared after i cleaned the scanner unit without disassembled it.

Thanks for any help
Best regards
Sara.ahmed.hazem@gmail.com

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Agfa cr 35

August 09, 2016 09:58

normally it is caused by dust of the optics

Log inor Register

to rate and post a comment

(34)
(2)

sara ahmed

re: re: Agfa cr 35

August 10, 2016 03:31

Dear mr.Lenny
Thanks for replay.
Please i want user name &password to enter service mode,can you help me ?
Best regards
Sara.ahmed.hazem@gmail.com

Log inor Register

to rate and post a comment

(7)

Soji Akinmosin

re: Agfa cr 35

August 09, 2016 10:05

There might be a particle along the slow scan direction. You may need to open it up and carefully clean it.

Log inor Register

to rate and post a comment

(34)
(2)

sara ahmed

re: re: Agfa cr 35

August 09, 2016 02:18

Thanks for all
Mr soji do you mean assemble scanner unit?how can i do it? I have service manual but it is not clear.
Best regards

Log inor Register

to rate and post a comment

(53)
(9)

EFREN FRANCO

re: re: re: Agfa cr 35

August 09, 2016 04:46

There is an internal brush that you can handle with a steel cable located at lower front of the equipment, just push and pull the cable two or three times and normally this corrects it.

Log inor Register

to rate and post a comment

(34)
(2)

sara ahmed

re: re: re: re: Agfa cr 35

August 10, 2016 03:31

Dear mr.Efren
Thanks for replay.
I use it but it is the same.
So i want to disassemble the scaner unit, can you help me?
Best regards
Sara.ahmed.hazem@gmail.com

Log inor Register

to rate and post a comment

(4)

Marty Bennett

re: Agfa cr 35

August 15, 2016 08:36

this is normally caused by the IP stalling while being scanned. You may have dislocated the backup roller in the scan unit. Very easy to do if cleaning the rollers without removing them. the backup roller is held only by clips.

Log inor Register

to rate and post a comment

(38)
(5)

Duane Ament

re: re: Agfa cr 35

September 23, 2016 09:34

Marty, I have an Agfa type 5175/220 with lines in the images, can you help me with this? Thank you! Duane.

Log inor Register