Ampronix, a Top Master Distributor for Sony Medical, provides Sales, Service & Exchanges for Sony Surgical Displays, Printers, & More. Rely on Us for Expert Support Tailored to Your Needs. Email info@ampronix.com or Call 949-273-8000 for Premier Pricing.
LifeImaging
Agfa CR 35-X error E 229F5
December 01, 2012 07:26
Service Engineers,
Actually I am looking for technical support for the following problem
on an Agfa CR-35 .
The AGFA CR-35X is jamming the Image Plates (both: 14x17 & 10x12). This
jamming broke one 10x12 image plate already.
Today, December 1st, 2012 we received a call because the CR had jammed
a 14x17. When we opened the side door, the 14x17 was stuck and there
was another 10x12 hidden stuck behind the cPCI-Rack.
We checked the errors in the maintenance menu and we have many
different errors. This errors are: 24640, 247A4, 247A6, 24305, 247A7,
22570 & 229F5.
Following the maintenance guidelines in the service manual, we gave a
full maintenance to the CR today, followed by multiple scanning tests.
We had 2 failures in 5 different scanning processes: first it "jammed"
the IP, apparently the rolls were unable to move the plate, even
though the plate was not jammed or bended and was removed easily;
second, the cassette did not exit the front slot of the CR correctly
and it fell inside the CR.
Now, the CR fails the initial self-test and shows the error 229F5,
which says that there is a problem with the "slow scan motor", this
error doesn t allow us to continue making tests .
The 2 most frequent errors are 247A6 & 229F5. These errors are
explained as follow:
247A6
Meaning: IP jam in scanner.
Reason: IP jam occurred while transport of IP from scanner position
bar down to light barrier in scanner. Light barrier remains free.
Cure: Open right side cover and put IP into cassette.
229F5
Reason: Slowscan
Module: Motor test of stepper node lift scanrolls failed.
Meaning: Motor not connected or defective.
Cure: Connect motor or replace.
Your help and input will be greatly appreciated.
Thanks in advance,
--
email: lifeimaging@hotmail.com
to rate and post a comment
Lenny Place
re: Agfa CR 35-X error E 229F5
December 03, 2012 10:09
Have you verified your belt tensions? If the belts become worn or streched it does not allow the rollers to put enough contact tension on the IP causing it to slip during transport.
to rate and post a comment
ALEXANDER PILAFTSIS
re: re: Agfa CR 35-X error E 229F5
February 25, 2016 11:25
I have an error 24794 with a CR agfa 35x. Do you know what might be the problem?
to rate and post a comment
Lenny Place
re: re: re: Agfa CR 35-X error E 229F5
February 25, 2016 03:59
Error Code --> 24794
E_HDCV_UIF_ASK_IDENTIFICATION,
Reason : Handling control module : Error number connected to user interface
confirm text for identification error of cassette to input.
Meaning : User fault of cassette identification.
Cure : Remove cassette and do identification.
buffer with correct orientation.
to rate and post a comment
Ronald Swapp
re: re: Agfa CR 35-X error E 229F5
January 12, 2018 09:38
There is no belt tension on a CR25/35. It's driven in by a cam in the Scan engine.
to rate and post a comment
Rebel Cook
re: Agfa CR 35-X error E 229F5
December 03, 2012 10:09
For the 229F5 error check the BU623 from ST3 to M621, reseat the interconnect.There is a service bulletin out for the 247A4 error that might help.The 22570 could be caused by a bad fuse on the power integration board. Hope some o9f this might be helpful.
to rate and post a comment
Reynaldo Rosa
re: Agfa CR 35-X error E 229F5
December 05, 2012 04:23
Thank you all for trying to help us solve this issue.
The system check log shows that there is a problem in: : M401-ROLLER-LIFT-MOT: motor connected ............................. failed .... (conn. to node GS410-SCAN-ROLLS-LIFT)
Here is the full System Check Log File:
SYSTEM CHECK | CR 35-X S/N 2276
------------ | 2012-12-05 13:36:10
General Information:
--------------------
Machine Name: CR25.0
Type: 5158
Subtype: 100
Software Version: c25_3103
CPU:
----
: "CPU": temperature check .......................................... ok
PCI AND IOBUS CHECKS:
---------------------
: "GS310-SCAN-CTRL-M-BD": existence ................................. ok
: "IOBUS": existence ................................................ ok
IOBUS NODES:
------------
: "IOBUS": general check ............................................ ok
: ADDR 0x21 GS710-5FOLD-STEP-BD SUBNODE1: response ................. ok
: ADDR 0x22 GS710-5FOLD-STEP-BD SUBNODE2: response ................. ok
: ADDR 0x26 GS712-IP-ALIGNMENT-BD: response ........................ ok
: ADDR 0x27 GS602-IP-TRANSPORT-BD: response ........................ ok
: ADDR 0x36 GS412-SLOWSCAN-BD: response ............................ ok
: ADDR 0x37 GS410-SCAN-ROLLS-LIFT: response ........................ ok
: ADDR 0x70 GS230-POWER-INTEGR-BD: response ........................ ok
: ADDR 0x75 GS210-ERASURE-CTRL-BD: response ........................ ok
: ADDR 0x85 GS116-EMERGENCY-NODE: response ......................... ok
MOTORS AND LIGHT BARRIERS:
--------------------------
: "Stepper Voltage": range check .................................... ok
: M701-CASS-ROLLER-MOT: power connected ............................. ok
: M701-CASS-ROLLER-MOT: motor connected ............................. ok
: GS734/GS732-CASS-IN-DET-LS: light barrier (0) connected ........... skipped ..
.. (always connected because LB driver is integrated on GS710!)
: ditto: light barrier (0) status .............................. clear
: GS726-CASS-POS_DET_LS: light barrier (X) connected ................ ok
: ditto: light barrier (X) status .............................. clear
: M702-CASS-OPENER-MOT: power connected ............................. ok
: M702-CASS-OPENER-MOT: motor connected ............................. ok
: GS716-CASS-OPEN-0-LS: light barrier (0) connected ................. ok
: ditto: light barrier (0) status ........................ interrupted
: GS720-CASS-SWITCH: light barrier (X) connected .................... ok
: ditto: light barrier (X) status .............................. clear
: M703-CASS-LIFT-MOT: power connected ............................... ok
: M703-CASS-LIFT-MOT: motor connected ............................... ok
: GS718-CASS-LIFT-0-LS: light barrier (0) connected ................. ok
: ditto: light barrier (0) status .............................. clear
: M704-CASS-BELT-MOT: power connected ............................... ok
: M704-CASS-BELT-MOT: motor connected ............................... ok
: GS714-CASS-BELT-0-LS: light barrier (0) connected ................. ok
: ditto: light barrier (0) status ........................ interrupted
: M705-CASS-CLAMP-MOT: power connected .............................. ok
: M705-CASS-CLAMP-MOT: motor connected .............................. ok
: GS722-CASS-CLAMP-0-LS: light barrier (0) connected ................ ok
: ditto: light barrier (0) status .............................. clear
: GS724-CASS-CLAMP-LS: light barrier (X) connected .................. ok
: ditto: light barrier (X) status .............................. clear
: M706-IP-ALIGNMENT-MOT: power connected ............................ ok
: M706-IP-ALIGNMENT-MOT: motor connected ............................ ok
: GS728-ALIGN-0POS-LS: light barrier (0) connected .................. ok
: ditto: light barrier (0) status .............................. clear
: GS730-IP-POS-DET-LS: light barrier (X) connected .................. ok
: ditto: light barrier (X) status .............................. clear
: M601-IP-TRANSPORT-MOT: power connected ............................ ok
: M601-IP-TRANSPORT-MOT: motor connected ............................ ok
: GS604-IP-TRANSP-0-LS: light barrier (0) connected ................. ok
: ditto: light barrier (0) status .............................. clear
: M402-SLOWSCAN-MOTOR: power connected .............................. ok
: M402-SLOWSCAN-MOTOR: motor connected .............................. ok
: GS442/GS444-IP-ALIGNMENT-LS: light barrier (0) connected .......... ok
: ditto: light barrier (0) status .............................. clear
: GS446-SCAN-IP-DET-LS: light barrier (X) connected ................. ok
: ditto: light barrier (X) status .............................. clear
: M401-ROLLER-LIFT-MOT: power connected ............................. ok
: M401-ROLLER-LIFT-MOT: motor connected ............................. failed ..
.. (conn. to node GS410-SCAN-ROLLS-LIFT)
: GS448-ROLLER-LIFT-LS: light barrier (0) connected ................. ok
: ditto: light barrier (0) status .............................. clear
: GS442/GS444-IP-ALIGNMENT-LS: light barrier (X) connected .......... ok
: ditto: light barrier (X) status .............................. clear
POWER:
------
: "24 V on Erasure Control Board": check ............................ ok
: "Power Unit": temperature check ................................... ok
: "Line Power": detect .............................................. ok
: "Auto Line Selection Stepper/Erasure-Transformers": check ......... ok
ERASURE:
--------
: "Erasure Fan": detect ............................................. ok
OPTICS:
-------
: "Scanner": entering service mode .................................. ok
: "Laser Control Board": detect ..................................... ok
: "Voltages": check ................................................. ok
: "Laser": test ..................................................... ok
: "Laser Power": test ............................................... ok
: "Polygon connected": check ........................................ ok
: "Scanner": leaving service mode ................................... ok
TAG READER:
-----------
: "Mifare Tag Reader": response ..................................... ok
ETHERNET / TCP/IP:
------------------
: "CR35 (192.200.0.201)": ping ...................................... ok
: "QSSERVER (192.200.0.200)": ping .................................. ok
: "QSSERVER-FP (192.200.0.200)": ping ............................... ok
: "QSSERVER-ID (192.200.0.200)": ping ............................... ok
- - - END OF FILE - - -
to rate and post a comment
LifeImaging
re: Agfa CR 35-X error E 229F5
December 11, 2012 07:15
Hello Engineers,
Service Code 229F5 was resolved by replacing the M401 Roller Lift Motor and also the Stepper Motor Board GS410. First we replace the M401 Roller Lift Motor because we found that the bearings of this part where damaged. As we keep getting the same error after replaced, we start to invert the GS410 Step.. Motor Board with the GS412 Stepper Board that are the same with different Jumpers configurations and we note that the error code given by the machine was different. We make same exercise again inverting the GS410 Stepper Mot. board now with the GS 602 IP Transport Board and we get another service code related with this section. Conclusion: "Stepper Board Damaged" . After the Stepper Motor Board GS410 replacement, machine is working properly without problems.
Thanks for all your support Friends!
to rate and post a comment
Salvatore Tovarich
re: re: Agfa CR 35-X error E 229F5
November 23, 2017 05:28
Hi could you help me with the error 22497 Agfa 25.0 thank you very much for your help
Regards
Tovarich
to rate and post a comment
Ronald Swapp
re: Agfa CR 35-X error E 229F5
January 12, 2018 09:38
Replace the suction cups.
Also replace the vac inline filters, the o-rings will fail with time causing a vac leak when the suction cups try to attach to the plate.
Another suggestion is to clean the back side of the IP, ONLY USE AGFA CR Plate cleaner.
A problem with the slow scan motor is usually a problem with that stepper card.
Remove the stepper card and move it to another location, be sure to readress it. If the error moves with the stepper card you'll have no choice but to replace it.
The stepper card is a common failure point in the CR25, 35, and 75 and 85's.