Agfa CR 35-X Error

Tags: Radiology - R/F, AGFA

(227)
(52)
(1)

Amos Dantene

Agfa CR 35-X Error

April 03, 2014 11:06

I cannot copy images from the Agfa CR 35-X cassette onto the computer because of the error:
"QSSERVER NOT READY"
with the code 216A7


Please you I also need the service manual for Agfa CR 35-X

Please my email address is: adantene@gmail.com

Thank You

1 person liked this
Log inor Register

to rate and post a comment

(58)
(18)

Rebel Cook

re: Agfa CR 35-X Error

April 03, 2014 06:10

This error is caused by the CR 35 not being able to communicate with the workstation. Usually it is a newtorking problem IE network is down, ip address has changed, something of this nature. After you have checked the network issue out restart the workstation and then the CR 35 and any images in the CR que will be sent to the worstation as long as the cr and worstation can communicate.

Log inor Register

to rate and post a comment

Mustafa RUZGAR

re: re: Agfa CR 35-X Error

April 04, 2014 08:16

PLEASE! Agfa medical devices and related services in your hand all my e-mail address to send documents

Log inor Register

to rate and post a comment

(42)
(4)

salman khan

re: Agfa CR 35-X Error

April 03, 2014 06:11

Hello Amos


This means having no communication with QS SERVER workstaion i think bad configuration or check your cable going to network hub if u can ping then bad configuration you will need CPF and xml files for configuration

Log inor Register

to rate and post a comment

(315)
(43)
(1)

Ronald Swapp

re: re: Agfa CR 35-X Error

June 21, 2018 10:51

Pinging the workstation or the cr does not test it's dicom handshake, you must do a dicom echo to test.

Log inor Register

to rate and post a comment

Mustafa RUZGAR

re: Agfa CR 35-X Error

April 04, 2014 08:16

PLEASE! Agfa medical devices and related services in your hand all my e-mail address to send documents

Log inor Register

to rate and post a comment

kad ahmed

re: re: Agfa CR 35-X Error

April 04, 2014 02:13

kaddour717@yahoo.fr , this is my email, ill be thankfull if u can help me on cr35 issues and also drystar 5503, im not able to calibrate film now , and only one try from the tree is working , ill be thankfull if u help me also with the nx configuration.

Log inor Register

to rate and post a comment

(23)
(3)

Luis Espinoza

re: re: re: Agfa CR 35-X Error

August 05, 2016 07:30

Email me: luismanuelespinoza@hotmail.es

Log inor Register

to rate and post a comment

AMMAR ALZUHAIR

re: Agfa CR 35-X Error

July 26, 2016 10:10

Hi good day all!
engineers i have got a really trick issue with my CR35-x, when i insert cassettes ( various sizes ) some working well however, some when it must get out from the CR itself it remain stuck inside, i meant the door doesn't open when it should open to allow the cassette to get out.

my email for any: ammaralzuhair@gmail.com
your cooperation is highly appreciated.

Log inor Register

to rate and post a comment

(2)

Iqbal Khan

re: Agfa CR 35-X Error

June 18, 2018 04:32

agfa cr 25.0 ERROR CODE 216A7 QSBC1 NOT READT

Log inor Register

to rate and post a comment

(432)
(83)
(3)

juan pablo amado

re: re: Agfa CR 35-X Error

June 19, 2018 08:06


216A7
E_RED_HOSTNOTAVAILABLE
Reason : The desired host is currently not available
Meaning : The host might be powered down, might not be able
to respond due to a lack of ressources, ...
Cure : Check the specified host - eventually reset its
background processes.


By I know:
Check LAN cable and destination config in cpf

Log inor Register

to rate and post a comment

(315)
(43)
(1)

Ronald Swapp

re: Agfa CR 35-X Error

June 21, 2018 10:51

Open either QS Server or NX Workstation and verify the Digitizer configuration as well as the workstation configuration.
Check the local IP Address, it must be static.
Verify the AE Title of the QS Server and or NX workstation that the CR is trying to communicate with.
Grab a fresh CPF from the digitizer and open it with the CCM tool and verify that the CPF configuration matches the QS Server or NX Workstation Configuration.
You can try to ping the digitizer from the QS Server or NX Workstation if you get a ping that does not necessarily mean that the two devices will communicate.
If you get a ping you must test a Dicom Echo to the digitizer. You can either do this from within the QS or NX Configurator or use a program called echoscu to test the handshake.
You'll probably find that A) the IP address of the Acquisition has changed, B) you have a bad network switch, C) System is on a managed switch and the mac address of the two are not in the switch management tables, D) your AE titles are incorrect either in the QS / NX Configurator or CR CPF file.

Anything further I can be reached at rtswapp@comcast.net

1 person liked this
Log inor Register

to rate and post a comment

(16)
(1)

karen

re: re: Agfa CR 35-X Error

January 14, 2019 09:41

i have software ,
I have the two software of the afga 35x digitizer on the computer and the service provider can send it on disk and cd. but I need the steps to install the program to the server. the equipment screen shows me a blinking point to start the start test

Log inor Register

to rate and post a comment

(24)
(2)

karen alberto garcia salazar

re: re: Agfa CR 35-X Error

January 14, 2019 09:41


I have the two software of the afga 35x digitizer on the computer and the service provider can send it on disk and cd. but I need the steps to install the program to the server. the equipment screen shows me a blinking point to start the start test
ing karen garcia
rymem dominicana
rymem2669@gmail.com

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Cbrx2sales
TIMED
Jun 05, 2026 (157 lots)