AGFA CR ADC compact Error 22D44
May 19, 2016 05:10
Hi there,
Could anyone help with this error 22D44. it's an ADC compact CR system from AGFA. any help will be more that welcome. Thanks
By the way, if anyone has the Error code book for ADC compact or ADC Solo CR system, I would certainly appreciate if he can e-mail it to me djamel@fte.ie
Thanks gents.
Regards
to rate and post a comment
Haider Ali
re: AGFA CR ADC compact Error 22D44
May 20, 2016 07:47
22D44
E_HDC_UIF_ASK_NOT_IDENTIFIED
Reason : Handling control module : Error number connected to user interface
confirm text for not identified cassette in input.
Meaning : User fault of cassette identification.
Cure : Remove cassette and do identification.
to rate and post a comment
Haider Ali
re: AGFA CR ADC compact Error 22D44
May 20, 2016 07:49
22D44
E_HDC_UIF_ASK_NOT_IDENTIFIED
Reason : Handling control module : Error number connected to user interface
confirm text for not identified cassette in input.
Meaning : User fault of cassette identification.
Cure : Remove cassette and do identification.
to rate and post a comment
Djamel Boutemine
re: re: AGFA CR ADC compact Error 22D44
May 20, 2016 11:52
Thanks gents...
Solved...
to rate and post a comment
juan pablo amado
re: AGFA CR ADC compact Error 22D44
May 20, 2016 07:48
22D44
E_HDC_UIF_ASK_NOT_IDENTIFIED
Reason : Handling control module : Error number connected to user interface
confirm text for not identified cassette in input.
Meaning : User fault of cassette identification.
Cure : Remove cassette and do identification.