I am looking for schematics of f8.5146.1430.x Power supply. It auto shut off immediatly after power on ( with 220v input). No problems with consumption,the fault is located in the 110/220V automatic range selection. At 110v no shut off but neither power on

Log inor Register

to rate and post a comment

(4)

Tech Nik

re: Agfa CR25 PS

November 16, 2013 08:09

I can send the CR-25.0-Chapter 02 Functional Description 2.0 and other manuals in exchange for Konica manuals.

An internal wiring diagram you will not find