Thomas Stenske

AGFA CR25 Solo , error code 24640

July 02, 2009 11:11

Intermittent error code 24630 I.P jam. No cassette is loaded and no plate is in the machine. Won't pass the self test without the same error. Any suggestions appreciated.

Log inor Register

to rate and post a comment

DJIELA Flavien

Re: AGFA CR25 Solo , error code 24640

July 20, 2009 07:13

please, carefully check the configuration on the workstation

Log inor Register

to rate and post a comment

(55)
(14)
(1)

Bao Vo Quoc

Re: AGFA CR25 Solo , error code 24640

October 30, 2009 02:04

Hi My Guy,
according to your description, I think that your CR25 will be failed as 1 in 2 case below:
1. The Pin Diode Board had failed (75%)
2. IP center calibration sensor failed(25%)

These part is inside the Laser Module so it is difficult for you to check and open it.

We need to re-check them again before decide to buy the new spare part.

The Best way you can contact your Agfa local for getting support in this case.

dont hesitate contact me if you have any question

Thanks and best Regards

Bao Vo Quoc
Field Service Engineer,AGFA HEALTHCARE VIETNAM
M: +84 903 901 698
E: baoagfa@vnn.vn or quocbao.vo@agfa.com




Log inor Register

to rate and post a comment