bashir ahmad
Agfa CR25
June 14, 2017 08:20
Hi friends I have a AGfa CR 25 which gives error 24793 with cassette size 43x35cm, and show wrong cassette, It works fine with 24x30cm cassette.
I have service manual but their I found nothing about this error. My email is bashir30474@yahoo.com
Regards
to rate and post a comment
juan pablo amado
re: Agfa CR25
June 14, 2017 02:39
24793
E_HDCV_UIF_ASK_WRONG_CASS
Reason : Handling control module : Error number connected to user interface
confirm text for wrong or wrong positioned cassette in input buffer.
Meaning : User fault Input of cassette in input buffer.
Cure : Remove cassette and put into buffer with correct orientation.
PD. Note:
Check for silver dot in cass margin
to rate and post a comment
RODRIGO CORTeS
re: Agfa CR25
June 15, 2017 01:43
Only have once 35x43? If only do It with one. Try to inicializate O replace RF card